Hi, hoping someone can help. I have acquired a Pintek PS-200 20MHZ oscilloscope and am attempting to calibrate it. So far, everything has been going fine - PSU voltages check out, brightness & focus fine, DC Step balance, DC Balance and 5x magnification, B-channel invert and Add(A+B) DC-gain all set Ok. Input capacitances and phase compensation all set, 100MHZ square wave all set Ok too. Speeds faster than around 10ms are all OK too. Square waves and sine waves above this speed look good, with correct vertical and horizontal displacement. So far so good.
But for speeds slower than 10ms , the trace (dot) slows down as it moves across the screen, and at the 2-second speed it actually stops at around 7/8ths of the screen and sometimes even moves back very slightly.
The sawtooth ramp off the sweep generator looks straight at speeds faster than 10ms, but curves off at slower speeds as per the attached image; the curve becoming flatter the slower the speed.
![[FIXED] Pintek Oscilloscope sweep slows down at slow timebases](https://www.eevblog.com/forum/index.php?action=dlattach;topic=297476.0;attach=1314950;image)
I've removed the sweep generator board and checked all diodes, resistances and capacitances, replacing a couple of electrolytics with no effect. I could really use help from someone more expert than I as to how to diagnose the issue further and what to check for,
Schematic of the sweep generator also attached.
11_b_PS 20-60mhz_compressed.pdf (26.55 kB - downloaded 42 times.)
Thanks,
Jon.