| daisizhou:
--- Quote from: Harry_22 on June 12, 2024, 01:00:28 pm ---Think it is Ok. You should not make any adjustments as this requires more precise measurement instruments than 8920A. --- End quote --- Got it, I'm not going to make any adjustments, I mean those adjustable resistors, adjustable capacitors. |
| daisizhou:
Based on yesterday's maintenance experience of cleaning P102, do P101 and P103 also need to be disassembled and cleaned like yesterday? Because I saw P103 has a "LO SENSE" mark |
| Harry_22:
Please check it again with 50 Ohm load at the input. |
| daisizhou:
--- Quote from: Harry_22 on June 13, 2024, 11:17:53 am ---Please check it again with 50 Ohm load at the input. --- End quote --- The input end is connected to a 50 ohm load, and the result is still the same. |
| Harry_22:
Ok, make an ajustment according paragraph 4-48 on page 4-19 steps from 1 to 8. Adjust only R72 according step 5 and R26 - step 8. |