Author Topic: Power Supply for Tektronix DPO2000B Series Scope  (Read 1249 times)
Offline paschulke2Topic starter
« on: August 17, 2017, 07:36:06 am »
Does anybody know whether the power supply modules for the DPO2000 and the DPO2000B series oscilloscopes are the same? I managed to find the service manual for the 2000 (without B) series and from the looks I'd say they are the same (part number 119-7266-00).
Just in case you want to ask: "Why don't you ask Tektronix?" I did. The gave me the part number for the mains cable ...
Thomas
Re: Power Supply for Tektronix DPO2000B Series Scope
« Reply #1 on: August 18, 2017, 01:49:10 pm »
Now I've received a second answer from Tektronix: The power supply for the DPO2000B series is not available at all, not even the 119-7266-00 for the DPO2000 series. They would not tell me if the power supply in the B models is the same. Note that the DPO2024B is still a current model.
I've just found out that I still have 8 months of warranty. I thought I had bought it more than five years ago ...
Thomas