Author Topic: Problems with built in batteries of a Tektronix 11801 Sampling Oscilloscope  (Read 3127 times)
Offline PhysikfanTopic starter
I have a Tektronix 11801 sampling oscilloscope, which does not start after switching on because of problems with built in batteries.
Please, are there experiences with such problems within this forum?
Online tggzzz
Re: Problems with built in batteries of a Tektronix 11801 Sampling Oscilloscope
« Reply #1 on: September 13, 2017, 09:59:03 am »
Which batteries?  What makes you think it is the batteries? Have you looked at the operator and service manuals? What have you found by googling?
The more information you give, the more likely you are to find someone that can help.
Also, try the Yahoo TekScopes group.
« Reply #2 on: September 18, 2017, 10:40:48 am »
Hi tggzzz
If you get error messages like this E5622, should you change the two Dallas NVRAMs against new ones?
Please these are my questions:
1. Is somebody here who has opened already an oscillocope like the Tek 11801C?
2. Is somebody here who has changed already the two Dallas NVRAMs against new ones?
3. And please, could I get pictures of these NVRAMs?
The service manual Tek 11801C unfortunately only gives this information:
4. Does somebody have more information about these special ICs?
Best regards
Physikfan
« Last Edit: September 18, 2017, 11:34:40 am by Physikfan »
« Reply #3 on: September 19, 2017, 10:49:01 am »
Hi
I got the information of an existing file:
"11801 and CSA803_TB_error 5622 repair.zip"
Which could possibly help to solve the problem.
This file should be available:
"You can access this file at the URL:
http://groups.yahoo.com/group/TekScopes/files/11801%20and%20CSA803_TB_error%205622%20repair.zip"
Unfortunately I have no idea how to get access to this file.
Please, could somebody help me?
Regards
Physikfan
Online tggzzz
« Reply #5 on: September 19, 2017, 05:18:36 pm »
Hi tggzzz
Thanks for your answer.
"https://groups.yahoo.com/neo/groups/TekScopes/info
Click "+ Join Groups"
Unfortunately I do not get the "+ Join Groups" button
at the page
"https://groups.yahoo.com/neo/groups/TekScopes/info"
Regards
Physikfan
Online tggzzz
« Reply #6 on: September 19, 2017, 05:57:09 pm »
Hi tggzzz
at the page
Regards
Physikfan
You must be using a very peculiar browser; it works in Firefox, Chromium and Opera.
« Reply #7 on: September 21, 2017, 10:02:31 am »
Hi tggzzz
Thank you for your comments regarding the error messages E5622 and E5322:
With the help of these two papers:
"Fixing a 50GHz Sampling Scope (Tektronix CSA803), posted by February 26, 2014 by xellers."
https://xellers.wordpress.com/2014/02/2%20...% 20ix-csa803 /
and
"Tek 11801 Timebase Errors by John Larkin, Highland Technology Inc, jjlarkin@highlandtechnology.com
11801 and CSA803_TB_error 5622 repair "
had a friend of mine (professional electronics engineer) and me after 3 hours work success with a Tek 11801A:
chipset
One of the error messages:
Successful repair:
Best regards
Physikfan
Online tggzzz
« Reply #8 on: September 21, 2017, 10:17:20 am »
Well done!