Hello. This is still a newcomer. I decided to ask you for help, as I myself couldn’t do anything yet.
I have an electronic load Akip-1301A 60V / 30A
(foreign analog of Prodigit 3320 60V / 30A) copy of 100%.
The malfunction lies in the inaccuracy of the readings that I submit to her voltmeter.
From 0 to 19 V is not resp., But from 20 to 60 V goes.
There are several tuning resistors, but so far they do not help.
On the Internet I was looking for a service manual or circuit, I did not find anything.
Maybe someone will help with the scheme or tell me where to go next.
-
Пост на всём кроме английского могут удалить, так что советую писать на нём. Тем более что большинство местного народа не будет заморачиваться переводить. И это не копия, а он и есть. Весь акип это тупо смена шильдика.
-
The auto-ranging of the meter may be broken.
-
One person suggested that it should be calibrated for accuracy. In the service manual it is. But on the Internet, I did not find it.
-
From 0 to 19 V is not resp., But from 20 to 60 V goes.
Do you mean up to 19V it does not work at all or you cannot adjust? If inaccuracy, how much. It could be a fault, not calibration problem.
-
measured the reading
-
(https://file01.itaiwantrade.com/e535185b-7aa9-47e7-8602-a02a4b41cdec/c56697a2-5da8-4eff-9711-8c929b1bcd87_3320_Specification.jpg)
So it should be accurate to +- 0.05 % + 2 counts, yours is up to 0.2 V out (15 V / 15.2 V), +1.3 % error.
Range switching is indeed done at 20 V. So the lower range needs adjustment, preferably to an reliable, accurate multimeter. Handheld 4 1/2 digit would probably be ok. There is probably one trimmer or 20-turn pot for each range.
-
I twisted some resistance, and put in place. They did not help. Only the general adjustment range was found
-
Hello everyone. Can I bring this topic up? Maybe someone got a service manual or calibration method for this device?