Thanks to another forum member, I've obtained a third unit! This one is slightly more unhealthy than the other two by the looks. The front panel is absolutely immaculate as it has been stored with the front protective case on. However the inside is not so good. Rather gunked up with various hairy things living in i.
After a quick earth check to make sure it wasn't going to kill me and a cursory inspection of its guts looking for burns and burst caps, a quick power up and drop into X-Y mode shows an absolutely huge shift horizontally. Step one as always is to check the power rails. First with a meter, then check any suspicious ones with a scope.
The meter check showed all the rails being within tolerance apart from the 110v rail which was down at 103v. So a quick poke with another unit:

Hmm that's not right. Quick poke of the unregulated line and it looks just as bad.

C4439, C4429 look to be the suspects here as there is precisely no smoothing on the unregulated line.
C4429 is specified as 1200uF 100v electrolytic. Replacing with 1200uF Nichicon 160v RS part 846-6694. £2.89 ex.
C4439 is specified as 550uF 100v electrolytic. Replacing with 560uF Rubycon 100v RS part 749-7235. £2.31 ex.
Ordered.