The 428 (A and B) is a fantastic and fascinating instrument, a true leap in capabilities when it was new. Today, one buys a Uni-T DC clamp for next to nothing, of course, but I've been mesmerised by the 428 as long as I've known about it.
I've got one, a late B with a glass fibre PCB. It most probably was a "hangar queen", because it had a manufacturing defect. One of the leads to one of the primary windings in the mains transformer was shorted to ground. It had wedged between the core and the chassis when the transformer was fitted. Apparently, the instrument survived QC (the defect was impossible to detect without disassembly) but must have failed in service.
It blew my mains fuse and the RCCB, and I removed the transformer to verify what was broken and there it was. Some shrink tubing around the damaged cable, and it is good as new.
Yes, the probe can be moved, but, as stated upthread, the instrument must be recalibrated to a new probe.
Congratulations, fellow 428 fan!