Author Topic: Quick repair project - Marconi 2024  (Read 33483 times)
Offline chicken
  • Rusty Coder
Re: Quick repair project - Marconi 2024
« Reply #50 on: October 26, 2017, 11:37:31 pm »
PS: The low voltage on AM Input leads me to a few additional theories (warning, I don't know anything about control loops or RF gear):
  • The RF level is the issue and the MCU can't pull AM Input any lower. But I haven't figured out yet, where the MCU would determine the absolute RF level.
  • The low AM Input causes the issue. This could due to
    a) a broken DAC, which I find unlikely as AM modulation works. Though there are multiple DAC outputs involved, maybe one just controls the RF level.
    b) a calibration issue, i.e. the "steady state" output of the DAC should be higher.
« Last Edit: October 26, 2017, 11:40:04 pm by chicken »
Offline TK
« Reply #51 on: October 27, 2017, 12:39:20 am »
@chicken, thank you for the detailed reply.  I changed a bunch of TL074C opams today, same result.  My unit PS measures +11V and -11.37V, so I suspect there is some sort of imbalance creating a lot of false.  I removed a 220uF cap after TIP42 (-11V) and the error goes away, output power is set at -2dBm (it used to be -137dBm)... no RF output activity, but no errors.  I will continue playing with the +11V and -11V supplies.
EDIT: I forgot to reconnect the cable from the CPU board to the RF board, that is why it had no RF output activity.  I am now getting a lot of 501 and 502 FRAC/N errors like the original post, but my control board does not have IC207, it has IC14 that is a MAX274BCWI (Active Filters 4th & 8th Order Continuous-Time ).  I am not sure if my board is newer or older than the one from OP.
2nd Update: the 501 / 502 FRAC/N errors go away when calibration 100 is executed.  Quote from Service Manual:
Pre-steer values are derived by an automatic calibration routine (Cal routine 100). If pre- steer problems are suspected, running this routine will cure the problem if no other hardware problems exist. If the routine cannot complete the calibration due to a hardware malfunction an error message is displayed to help identify the problem.
« Last Edit: October 27, 2017, 10:14:45 pm by TK »
Offline TK
« Reply #52 on: October 28, 2017, 12:29:50 pm »
I think I found an error on the Marconi 2024 schematics (Marconi_2023A_2024A_2025_Service_Manual.pdf, page 248, Level Modulator AA1 Module Schematics sheet 6 of 9).  The inputs to the opamp TL074C(a) are inverted.  I checked on the board and what was drawn as going into (-) pin 2 goes to (+) pin 3 and viceversa.
Offline chicken
  • Rusty Coder
« Reply #53 on: October 28, 2017, 05:23:25 pm »
Good catch! I can confirm that in my unit as well as Dave's teardown:
https://www.flickr.com/photos/eevblog/7004690115/in/album-72157629274319878/
(ST chip in the center)
Offline chicken
  • Rusty Coder
« Reply #54 on: October 28, 2017, 05:30:12 pm »
Actually, in the service manual that I have (2023 & 2024, (c) 1998), the schematic is correct.
Offline TK
« Reply #55 on: October 31, 2017, 12:02:41 pm »
@chicken: My manual is 2023A 2024A and 2025 service manual and it has many errors.  I cannot find the 2023 2024 service manual you have, did you get it online?
Found it
« Last Edit: October 31, 2017, 12:08:49 pm by TK »
Offline GB97816
« Reply #56 on: December 05, 2017, 05:37:43 pm »
Hello,
sorry that I put in here. Does anyone of you have a binary file of IC5 of IFR2024 (EPROM 27C4001) and could email me this?
In my device, the block seems to be broken, I see in the hex file destroyed pleadings.
Thanks and Regards,
Georg
Online gby
« Reply #57 on: July 29, 2018, 04:34:46 pm »
I just got a used Aeroflex/IFR/Marconi 2025 sig gen from fleabay.  It seemed to work at first, but after 5 days of being off I pushed the power button, the display lit up with blue background, and then no text ever. :'(  After lots of time/power cycling with still no display text I figured I had a repair project....
When I opened it up everything looked as expected and no sign/smell of burnt electronics.  Then I took the plate off the power supply section and got a big surprise.  There was not a single ac in to multiple voltage out supply board like in the similar 2023A EEVBlog tear down, 2024 pictures in this forum posts.  There were 4 individual metal boxed ac in to single dc voltage out supplies crammed into there! :scared:  So, someone had done a complete home brew power supply replacement.  See Picture.
I then checked the power supply voltages and got +24V, +12V, -12V as expected but the +5V was only 2.45 V.  Even worse, the current was from the device into the +5V supply (backwards flowing).  So, the 5V was gone.  Since the 5V supply runs the processors it makes sense that there was no text.
Next step is to get a working 5V supply and see if that brings the unit back to life.  The existing substitute supply was labeled 5V, 3.8 A.
My question to the group is how much 5V current does this unit need/use??
I could not find any specs in the Service Manual nor written on the original power supply PCB pictures I found from tear downs.
Could someone take a look at the supply board carefully and see if any load current capacities are on the silk screen or if you could hook a current meter like the UT210E around the 3 orange wires and measure the load current while operating?
Thanks.
Online PA0PBZ
« Reply #58 on: July 29, 2018, 07:09:44 pm »
Nice surprise power supply you have there! Can't you hook up a bench PSU to see if the thing will work and to find out the current consumption?
Offline Andy Watson
« Reply #59 on: July 29, 2018, 07:35:52 pm »
Table at the bottom of 1-39, in the serevice manual lists the "instrument requirements" of:
+5V  2.8A
+12V  2.3A
-12V  0.6A
+24V 0.6A
The following users thanked this post: gby
Online gby
« Reply #60 on: July 29, 2018, 09:00:16 pm »
Wow, I must have skimmed around the service manual for quite a while and could not find the current ratings.  So, thanks Andy for the pointer.  In my service manual pdf the supply chart is on page 1-40.  Who knew the instrument requirements would be in a table documenting the optional DC power version.
PA0PBZ I thought of hooking up a bench supply to continue the debug and finding the consumption.  But, I am worried about power sequencing when turning on.  Some multi-supply systems like the supplies turned on in a certain order and/or within a certain amount of time.  Of course, my unit's home brew supply probably lost all supply order sequencing compared to the original...
I will update this thread when I get time to spend more time on it.
Online gby
« Reply #61 on: September 09, 2018, 10:07:06 pm »
It took me a while to get a new supply since I ordered a replacement one from China.  When I checked the new supply outside the instrument I thought I would check the old one I took out.  Well the old one seemed to work.  1 month ago in the instrument it put no voltage out.  So, somehow the old one is flaky.
I put the new one in and the unit powered right up with the display showing numbers as it should.  I was able to set carrier frequency and amplitude.  For reference the 5V load was 1.9 A.  Certainly a weird/non-standard supply for this generator but at least it works again.
Now I have figure out the modulation modes and check those out.
Offline corvo
  • Country: pt
« Reply #62 on: February 22, 2019, 10:22:17 am »
This is a simmilar problem but for a different equipment series:
I'm looking for informations about one Display for the equipment:
 MARCONI INSTRUMENTS 10KHz - 1.35GHz signal generator 2030 (The Serial Number is 119329/044 Made in England).
The Display is:
 EPSON ECM-A0310
 (9006C0U)
Assuming it is a discontinued part:
Where can I find the data sheet?
Has anyone found a direct replacement?
Has anyone adapted a diferent display to this design?
Has anyone added new hardware to adapt a new display for this generator?
Running the test procedure it shows up the message:
**Can Not Setup Link Between Processors**
and the display shows some corrupted irregular half columns and lines as well.
I have checked all connections from the processor board to the display controller board and it matches as supposed.
 The controller board has the controller IC SED1330FBA
https://www.lcd-module.de/eng/pdf/zubehoer/sed1330.pdf
The pinout of the display is:
PLFM
1 plus 5 V
2 GROUND
3 VLCO
4 LP
5 FR
6 NC
7 NC
8 DIN
9 MSXCL
10 NC
11 D0
12 D1
13 D2
14 D3
I imagine a product: Display Box Interface, to convert obsolete displays to new ones.
If it doesn't exist, maybe with the information for the existing parts, we can build one.
Here is another place of information:
https://www.vintage-radio.net/forum/showthread.php?t=153435
https://groups.yahoo.com/neo/groups/marconi_test_instruments/conversations/topics/6244
Thank you.