| Dexter2:
Just re-burned my EPROM. Works without problems no CRC checks :clap:. Only one thing to watch out. To calibrate voltage you need a high voltage probe I used a HAMEG 50kV high voltage probe connected to a calibrated 10Meg input resistance DMM. The problem is as it seems these probes are not to precise, so before doing calibration determine the exact factor for the high voltage probe dont take the 1:1000 for granted it could be quite some % off. |
| alm:
Some DMMs have some variation in their input resistance (some even go to 10 GOhm). Also, I imagine a 50 kV probe might be adjusted for the end of its range (like the SR calibration, apparently ;)). Due to voltage coefficients, the division ratio may not be all that flat across its voltage range. I can see how that would make checking linearity of the power supply tricky without another HV source. |
| Dexter2:
--- Quote from: smgvbest on August 01, 2017, 10:55:31 am --- 1. Continue looking at code to verify all calibration options. 2. Verify calibration options (testing at DUT) 3. Develop and test manual calibration procedure (for VSET looks like this may be done) 4. Develop and test automated calibration procedure (python or perl code preferred as not everyone has other tools) --- End quote --- For me getting the calibration done is enaugh im happy. But would be interested in the unknown locations that we have not jet uncovered. I know my calibration assist program is not in a free lauguage but it was the quickest i could make one as i use LabView all the time. I think National Instrument also has a LabView 45day free full working trial if you want to play with it. If required i can make an executable you would then only need to download a free runtime engine to get it working. To automate the software more it would be possible to read in the offset and the two factors, read values from a external DMM in all three calibration points for all parameters and then calculate the new factors but it probablly does not make sense for a home calibration. |
| Dexter2:
Just a quick note as i did not write any instructions for the calibration program. First you calibrate all the wanted factors then you read out the calibration memory with "Read calibration memory". Next open your original BIN of the whole EPROM content and than just press write new EPROM. The software generates the new BIN ready to be written into the EPROM. |
| Dexter2:
Updated the table alm created with locations for readback calibration below. --- Quote ---WORD? # VALUE ROMHEX 0 2476 AC09 ; +pol vset offset 1 21988 E455 ; +pol vset exponent/multiplier/non-linear parameter if vset < word0 2 21003 0B52 ; +pol vset exponent/multiplier/non-linear parameter if vset >= word0 3 2500 C409 ; +pol vlim offset 4 21076 5452 ; +pol vlim exponent/multiplier/non-linear parameter if vlim < word3 5 21076 5452 ; +pol vlim exponent/multiplier/non-linear parameter if vlim >= word3 6 2492 BC09 ; +pol ilim offset 7 21017 1952 ; +pol ilim exponent/multiplier/non-linear parameter if ilim < word6 8 20051 534E ; +pol ilim exponent/multiplier/non-linear parameter if ilim >= word6 9 2491 BB09 ; +pol itrp offset 10 21032 2852 ; +pol itrp exponent/multiplier/non-linear parameter if itrp < word9 11 20053 554E ; +pol itrp exponent/multiplier/non-linear parameter if itrp >= word9 12 2476 AC09 ; +pol voltage readback display offset0 13 49126 E6BF ; +pol voltage readback display scale if VOUT < Word12 14 49634 E2C1 ; +pol voltage readback display scale if VOUT >= Word12 15 2493 BD09 ; +pol current readback display offset 16 49632 E0C1 ; +pol current readback display scale if IOUT < Word15 17 50148 E4C3 ; +pol current readback display scale if IOUT >= Word15 18 0 0000 19 17299 9343 ; **unknown** 20 17299 9343 ; **unknown** 21 2470 A609 ; -pol vset offset 22 21910 9655 ; -pol vset exponent/multiplier/non-linear parameter if vset < word21 23 20982 F651 ; -pol vset exponent/multiplier/non-linear parameter if vset >= word21 24 2500 C409 ; -pol vlim offset 25 21076 5452 ; -pol vlim exponent/multiplier/non-linear parameter if vlim < word24 26 21076 5452 ; -pol vlim exponent/multiplier/non-linear parameter if vlim >= word24 27 2513 D109 ; -pol ilim offset 28 20785 3151 ; -pol ilim exponent/multiplier/non-linear parameter if ilim < word27 29 19760 304D ; -pol ilim exponent/multiplier/non-linear parameter if ilim >= word27 30 2512 D009 ; -pol itrp offset 31 20818 5251 ; -pol itrp exponent/multiplier/non-linear parameter if itrp < word30 32 19746 224D ; -pol itrp exponent/multiplier/non-linear parameter if itrp >= word30 33 2470 A609 ; -pol voltage readback display offset 34 49170 12C0 ; -pol voltage readback display scale if |VOUT| < Word33 35 49634 E2C1 ; -pol voltage readback display scale if |VOUT| >= Word33 36 2514 D209 ; -pol current readback display offset 37 49784 78C2 ; -pol current readback display scale if |IOUT| < Word36 38 50308 84C4 ; -pol current readback display scale if |IOUT| >= Word36 39 0 0000 40 17299 9343 ; **unknown** 41 17299 9343 ; **unknown** 42 0 0000 43 51830 76CA ; **unknown** 44 0 0000 45 51830 76CA ; **unknown** --- End quote --- EPROM values are based on Sandra's EPROM and are helpfull for orientation. The locations first thought to be model specific that turned out to be calibration factors for VLIM are probablly even not calibrated in factory and are just preset with default values as they are all the same for all EPROMs. So if you want to have the unit preset to default initial values to calculate the calibration factors for voltage/current set/readback afterwards this probablly are the initial ones. Offset:2500 Scale:21076. |