alm:
The VLIM / ITRP values obviously do not need to be very accurate, it may very well be that they are the same within a certain model (although this does not seem to be the case for ITRP). This is the calibration data from a PS310 1.4 rom from the older thread:
--- Code: ---0. 6C 02
1. 1D 54
2. 56 53
3. 71 02
4. 54 52
5. 54 52
--- End code ---
Note the different offset (W3). The slope appears to be the same.
alm:
--- Quote from: alm on August 01, 2017, 12:50:27 pm ---Thanks! That makes sense. I have not looked at the readout display at all. Now that you mention it, I should search for the address of word 12 and will likely find the readout code.
--- End quote ---
I found the code. The function that does this starts at 0x0a39. I am still investigating the details, but similar to the settings I see three sets three words. Calculation with the three words looks fairly similar to the setting. I am not sure yet what the third set of three words refers to yet, I will have to look at the ADC code in more detail.
It involves a checksum (bytewise XOR) over a bunch of bitfields that are modified at various places. By toggling bits in the bitfields, it turns off bits in the checksum. The program relies on the checksum to tell which value to update. I need to identify all variables and trace where they are being updated and what they mean.
--- Quote from: alm on August 01, 2017, 12:50:27 pm ---There appears to be support for a multiplier for VSET (for lower voltage ranges) through 0x099f (I plan to check the PS310 firmware from the other thread to see if it has non-zero values as I would expect). [...]
--- End quote ---
I checked the PS310 1.4 (without GPIB) firmware from the previous thread Sandra linked to, and indeed the block of four bytes (located at a slightly different address) is non zero:
--- Code: ---ROM:095C unk_95C: db 2 ; DATA XREF: sub_898+53o
ROM:095D db 2
ROM:095E db 1
ROM:095F db 1
--- End code ---
This would scale the values sent to the DAC so the smaller voltage range would still correspond to full scale on the DAC. Though it appears that current values are also scaled up. I wonder if there is a separate block for readout, or that they just change the readout calibration constants to match? I am not sure if figuring this out is really a priority, but it at least it confirms my theory that the parameters at 0x99F (0x95C in the PS310 firmware) are model-specific.
--- Quote from: alm on August 01, 2017, 12:50:27 pm ---
--- Quote from: Dexter2 on August 01, 2017, 11:33:29 am ---And for the end the new memory map. Still a few locations are unknown. Any Ideas?
--- End quote ---
I will find the readout code and see if I can find a third readout calibration factor.
--- End quote ---
See above. Working on it.
Other thing I noticed. Sandra (and the schematic) has U507 Q4 as NC, but this code (that appears to shutdown the supply in case of I_FAULT, PRI_FAULT or V_FAULT), sets it:
--- Code: ---ROM:0D42 shutdown: ; CODE XREF: sub_D94+15p
ROM:0D42 ; sub_D94+57p ...
ROM:0D42 ld hl, port_78h_value
ROM:0D45 set 4, (hl) ; spare???
ROM:0D47 res 0, (hl) ; set -shutdown low
ROM:0D49 ld a, (port_78h_value)
ROM:0D4C out (78h), a ; set misc switches (u507)
--- End code ---
Old code? different model? Error in schematic?
Does anyone know what J504 (status1 bit 6) connects to? My copy of the schematics is not searchable, and a quick scan through the schematic pages did not reveal anything. There is a function at 0x108c that checks it and shuts down the power and sets the voltage to zero if it is high. Could be another fault or interlock indicator?
Does anyone know what port 0x3a is connected to? It is reading from it in a function that also writes to the display (for now named something_display :P):
--- Code: ---ROM:01C9 loc_1C9: ; CODE XREF: sub_15D+66j
ROM:01C9 in a, (3Ah)
ROM:01CB bit 7, a
ROM:01CD jp z, loc_227
ROM:01D0 set 1, c
ROM:01D2 jp loc_227
--- End code ---
Appears to be related to editing values on the display, so could be keyboard or display.
smgvbest:
here's the details on port 3Ah and the 20h bit 6 which is the rear panel switch to select between monitor mode or set mode for the V_MON/I_MON connectors on the back
this is my most current port and address map
--- Code: ---
MEMORY ADDRESS OF EPROM: 0000~1FFF (PROGRAM SPACE:8KB,ST M2764A)
SRAM: 4000~47FF (DATA SPACE:2KB,HITACHI HM6116)
IRQ_ROUTIINE: 0038h
SOURCE: SHEET 6
CLK->(U610)->TIMER @ 976.56hz/0.001024ms (4Mhz/4096(U610)74HC4020)
GPIB
SRAM ADDRESSES:
SRAM_MISC = 413AH
SRAM_STATUS1 = NA
SRAM_STATUS2 = 401CH
EPROM ADDRESSES:
*IDN? STRING 000019fah: 53 74 61 6E 66 6F 72 64 52 65 73 65 61 72 63 68 ; StanfordResearch
00001a0ah: 53 79 73 74 65 6D 73 2C 50 53 ; Systems,PS
191CH GPIB OFFSETS TABLE 0000191ch: 24 35 19 2A 3B 19 48 6E 19 49 79 19 53 89 19 54 ; $5.*;.Hn.Iy.S‰.T
CMD CHAR+OFFSET 0000192ch: 8F 19 56 9A 19 57 AA 19 00 ; .Vš.Wª..
193BH GPIB COMMAND 00001935h: 44 41 43 BE 1C 00 43 4C 53 6A 1A 45 53 45 3C 1E ; DAC¾..CLSj.ESE<.
CMD+OFFSET+00H 00001945h: 45 53 52 AE 1E 49 44 4E B0 19 50 53 43 72 1E 52 ; ESR®.IDN°.PSCr.R
00001955h: 43 4C 5A 1D 52 53 54 4E 1A 53 41 56 35 1D 53 52 ; CLZ.RSTN.SAV5.SR
00001965h: 45 06 1E 53 54 42 08 1F 00 56 4F 46 14 1A 56 4F ; E..STB...VOF..VO
00001975h: 4E 24 1A 00 4C 49 4D 7F 1B 4F 55 54 7E 1C 54 52 ; N$..LIM.OUT~.TR
00001985h: 50 EE 1B 00 4D 4F 44 9E 1C 00 43 4C 52 3B 1A 4D ; Pî..MODž..CLR;.M
00001995h: 4F 44 FF 1C 00 4C 49 4D 06 1B 4F 55 54 5D 1C 53 ; ODÿ..LIM..OUT].S
000019a5h: 45 54 87 1A 00 4F 52 44 ; ET‡..ORD
19FAH 000019fah: F3 CD D5 11 FB 0E 00 C3 D4 15 CD 9D 17 C2 FD 17 ; óÍÕ.û..ÃÔ.Í.Âý.
00001a0ah: 21 8D 1F 11 3C 41 01 09 00 ED B0 ; !..<A...í°
1F7BH 00001f7bh: 5A 31 33 34 ; Z134
1F7FH EPROM_VERSION_NUMBER 00001f7fh: 31 2E 33 34 ; 1.34
1F83H 00001f83h: 88 13 88 13 82 14 82 14 64 00 ; ˆ.ˆ.‚.‚.d.
1F8DH 00001f8dh: 00 00 88 13 82 14 82 14 00 00 00 00 00 00 00 00 ; ..ˆ.‚.‚.........
00001f9dh: 0E 48 0D ; .H.
1FA0H EPROM_CAL_DATA 00001fa0h: AC 09 E4 55 0B 52 C4 09 54 52 54 52 BC 09 19 52 ; ¬.äU.RÄ.TRTR¼..R
00001fb0h: 53 4E BB 09 28 52 55 4E AC 09 E6 BF E2 C1 BD 09 ; SN».(RUN¬.æ¿âÁ½.
00001fc0h: E0 C1 E4 C3 00 00 93 43 93 43 A6 09 96 55 F6 51 ; àÁäÃ..“C“C¦.–UöQ
00001fd0h: C4 09 54 52 54 52 D1 09 31 51 30 4D D0 09 52 51 ; Ä.TRTRÑ.1Q0MÐ.RQ
00001fe0h: 22 4D A6 09 12 C0 E2 C1 D2 09 78 C2 84 C4 00 00 ; "M¦..ÀâÁÒ.x„Ä..
00001ff0h: 93 43 93 43 00 00 76 CA 00 00 76 CA ; “C“C..vÊ..vÊ
1FFCH EPROM_SERIAL_NUMBER 00001ffch: 31 31 33 37 ; 1137
GPIB
DAC/$DAC CLS/*CLS ESE/*ESE ESR/*ESR
IDN/*IDN PSC/*PSC RCLZ/*RCL RSTN/*RST
SAV/*SAV SRE/*SRE STB/*STB VOF/HVOF
VON/HVON LIM/ILIM OUT/IOUT TRP/ITRP
MOD/SMOD CLR/TCLR MOD/TMOD LIM/VLIM
OUT/VOUT SET/VSET ORD/WORD
I/O ADDRESSES:
U504 (A3-A5 -> A-C, A6-G1, WR->!G2A, IORQ->!G2B )
40H: SPARE
48H:-SETS -> U407/U408
------------------------------
BIT0: U407A / GND->U405.3
BIT1: U407B / U410.RF -> U405.3
BIT2: U408A / ANALOG_VOLTAGE
BIT3: U408B / I_OUT_MONITOR
BIT4: U408C / CV_SET
BIT5: U408D / DV_LIM
BIT6: U407C / I_LIM
BIT7: U407D / I_TRIP
------------------------------
50H:-DAC U410 12BIT DAC AD7542
50H DAC_LOW_NIBBLE
51H DAC_MID_NIBBLE
52H DAC_HIGH_NIBBLE
53H DAC_LOAD
58H:-LED U611 74HC374
------------------------------
BIT0: L0
BIT1: L1
BIT2: L2
BIT3: -AD1/V_MON
BIT4: -AD2/I_MON
BIT5: -AD3/ANALOG_VOLTAGE_IN
BIT6: -IO
BIT7: UNUSED
------------------------------
60H:-SEGB U606 74HC374
------------------------------
BIT0: EDP
BIT1: EG
BIT2: EF
BIT3: EE
BIT4: ED
BIT5: EC
BIT6: EB
BIT7: EA
------------------------------
68H:-SEGA U609 74HC374
------------------------------
BIT0: ODP
BIT1: OG
BIT2: OF
BIT3: OE
BIT4: OD
BIT5: OC
BIT6: OB
BIT7: OA
------------------------------
70H:-STROBE (STROB0-STROB5 LED) (STROB2-4 KEYBOARD SCAN OUT)
------------------------------
BIT0: STROB0 (LED)
BIT1: STROB1 (LED)
BIT2: STROB2 (LED,KBD)
BIT3: STROB3 (LED,KBD)
BIT4: STROB4 (LED,KBD)
BIT5: STROB5 (LED)
BIT6: STROB6 ()
BIT7: STROB7 ()
------------------------------
78H: MISC -> (U507,74HC374 OCTAL D-TYPE FLIP-FLOP) >CLK
------------------------------
BIT0: -SHUTDOWN
BIT1: UPOK
BIT2: -FLAG_RESET
BIT3: -TIMER_RESET
BIT4: (SPARE)
BIT5: -POS
BIT6: -NEG
BIT7: FILTER
------------------------------
U505 (A3-A5 -> A-C, !M1->G1, A6->!G2A, IORQ->!G2B )
00H: SPARE
08H: SPARE
10H: SPARE
18H: SPARE
[b]20H: STATUS1 U506->74HC244
------------------------------
BIT0: COMPARE
BIT1: I_FAULT
BIT2: CUR_LIM
BIT3: PRI_FAULT
BIT4: V_FAULT
BIT5: TIMER
BIT6: -J504/SET_MONITOR_SW
BIT7: GND[/b]
------------------------------
28H: STATUS2 U512->74HC244
------------------------------
BIT0: HV_ON
BIT1: HV_OFF
BIT2: J501/-HV_SW_POS
BIT3: J502/-HV_SW_NEG
BIT4: GND
BIT5: GND
BIT6: GND
BIT7: GND
------------------------------
30H: -KBD
------------------------------
BIT0: KBD0
BIT1: KBD1
BIT2: KBD2
BIT3: KBD3
BIT4: KBD4
BIT5: KBD5
BIT6: KBD6
BIT7: KBD7
key press display
01 Man
02 GPIB
03 left-arrow
04 Right arrow
05 up-arrow
06 down-arrow
07 Enter
08 Select
09 STO
10 RCL
11 CLR
12 .
13 0
14 7
15 4
16 1
17 8
18 5
19 2
20 9
21 6
22 3
------------------------------
38H: -GPIB
------------------------------
READ
38H GPIB_INT_STATUS_0
BIT0: INT0 (1 WHEN STATUS 1 BIT SET)
BIT1: INT1 (1 WHEN BITS 2-7 OF STATUS 0 SET)
BIT2: BI (BYTE IN)
BIT3: BO (BYTE OUT)
BIT4: END (LAST BYTE RECEIVED)
BIT5: SPAS (RSV1/2 SERVICE REQUEST)
BIT6: RCL (REMOTE/LOCAL CHANGE)
BIT7: MAC (ADDRESS CHANGE)
39H GPIB_INT_STATUS_1
BIT0: GET (GROUP EXECUTE TRIGGER)
BIT1: ERR (ERROR)
BIT2: UNC (UNRECOGNIZED COMMAND)
BIT3: APT (ADDRESS PASS THROUGH)
BIT4: DCAS (DEVICE CLEAR ACTIVE STATE)
BIT5: MA (MY ADDRESS)
BIT6: SRQ (SERVICE REQUEST)
BIT7: IFC (INTERFACE CLEAR)
[b]3AH GPIB_ADDRESS_STATUS
BIT0: REM
BIT1: LLO
BIT2: ATN
BIT3: LPAS
BIT4: TPAS
BIT5: LADS
BIT6: TADS
BIT7: ULPA[/b]
3BH GPIB_BUS_STATUS
3EH GPIB_CMD_PASS_THRU
3FH GPIB_DATA_IN
WRITE
38H GPIB_INT_MASK_0
39H GPIB_INT_MASK_1
3BH GPIB_AUXILIARY_CMD
3CH GPIB_ADDRESS
3DH GPIB_SERIAL_POLL
3EH GPIB_PARALLEL_POLL
3FH GPIB_DATA_OUT
------------------------------
--- End code ---
Zucca:
Just want to say thank you, I am enjoying the show here.
I am impressed how the eevblog sometime sets up a wonderful skilled team to solve the EE problems.
:-+
smgvbest:
--- Quote from: Dexter2 on August 01, 2017, 01:53:41 pm ---I know my calibration assist program is not in a free language but it was the quickest i could make one as i use LabView all the time. I think National Instrument also has a LabView 45day free full working trial if you want to play with it. If required i can make an executable you would then only need to download a free runtime engine to get it working. To automate the software more it would be possible to read in the offset and the two factors, read values from a external DMM in all three calibration points for all parameters and then calculate the new factors but it probably does not make sense for a home calibration.
--- End quote ---
No problem and thank you for doing this so quick.
The trial is 7 days but it says you can extend 45 day inside the product. (yes I downloaded it :) )
I did play with the 1st three offsets and now my unit is <1V of target value full scale.
will play with rest as I have time.