I think you might want to add 0x99F (in your PS350 firmware) as four bytes that define a multiplier for the current/voltage scales, since I verified them to be different between the PS310 and PS350. This is how I described it in my listing:

--- Quote ---; Four byte array that defines exponents for scale adjustment for vset/vlim/ilim/itrp: input values are multiplied by 2^n after subtracting the offset but before applying the other calibration parameter.

--- End quote ---
This would mean the voltages were multiplied by four and the current by two to map to the DAC values according to the PS310 dump I looked at. This exactly matches the 1.25 kV vs 5 kV voltage range. I am guessing they use an 8x lower value shunt resistor to compensate for the increased current (20 mA / 8 * 2 = 5 mA).

Any idea what the relevance of ulpa (last LSB of address recognized by TMS9914, 0x3a bit 7) is? There is one function (at 0x674) that checks it (and some other variables) to decide to execute RTL through gpib_rtl (aux command 0x7):

--- Code: ---ROM:06C6                 in      a, (3Ah)
ROM:06C8                 bit     7, a
ROM:06CA                 jr      z, loc_6DA
ROM:06CC                 bit     6, a
ROM:06CE                 jr      z, loc_6D1
ROM:06D0                 ret
ROM:06D1 ; ---------------------------------------------------------------------------
ROM:06D1
ROM:06D1 loc_6D1:                                ; CODE XREF: gpib_something_rtl+5Aj
ROM:06D1                 ld      a, 3
ROM:06D3                 cp      b
ROM:06D4                 jr      nz, locret_6D9
ROM:06D6                 call    gpib_rtl

--- End code ---
The only thing I can come up with is a special GPIB address to configure it as talk-only, but I could find no mention of it in the manual. What other significance could the LSB have, since the address could be both odd and even? Or am I missing something about GPIB? I am reasonably familiar with using GPIB, but have never implemented a GPIB device.

I have now pretty much figured out the main loop. The IRQ handler at 0x38 will update the DACs, handle some GPIB and eventually jump back to it unless it encounters a (presumably fatal) error:

--- Code: ---ROM:00E3 main_loop:                              ; CODE XREF: sub_C3+1Bj
ROM:00E3                                         ; irq_service+C1j ...
ROM:00E3                 ei
ROM:00E4                 call    parse_gpib_commands
ROM:00E7                 call    gpib_something_rtl
ROM:00EA                 call    check_j504
ROM:00ED                 call    update_displays
ROM:00F0                 call    upd_settings_checksum
ROM:00F3                 call    do_readouts
ROM:00F6                 call    display_readout
ROM:00F9                 jp      main_loop

--- End code ---
Obviously the gpib_something_rtl is still unclear, which is where the snippet posted above came from.

smgvbest:
This is the definition of the ULPA bit

--- Quote ---This bit shows the LSB of the last address recognized by the TMS9914A
--- End quote ---

this is a bit further explaination of it

--- Quote ---The' edpa' bit is used to enable the dual addressing mode of the TMS9914A. It causes the LSB of the address to be
ignored by the address comparator giving two consecutive primary addresses for the device. The address by which
the TMS9914A was selected is indicated by the 'ulpa' bit of the Address Status Register.
--- End quote ---

ah,  I just noticed by the schematic they attached the D0-D7 to the TMS9914A in reverse   ie.  D0->D7, D7->D0 (see attached)
so if that is true then bit 7 is not ULPA but is REM
and REM is


--- Quote ---The device is in the remote state
--- End quote ---

alm:
Ah, that explains it! Checking REM before GTL makes complete sense. I guess they reversed the connection because the TMS9914A seems backward itself (D7 is the LSB). So now the MSB on the Z80 is also the MSB on the TMS9914A. I had actually already assumed that was the case (D0 connected to the LSB) without checking the schematic when decoding the command (GTL = 0x7).

smgvbest:
I need to update my portmap to flip these bits around to match.

smgvbest:
Here's my updated port/memory map with the GPIB BIT Order fliped

and just because I'm still looking at things.
And the switching frequency of the 555 timer driving the HV transformer is ~39.925Khz
this is fed into a flipflop then out to 2 3-Input NAND gates to drive the GATE of the 2 IFR532's
level is set via the VDRIVE signal to the COMMON of the transformer. The HVOFF signal disables the output of the 2 NAND gates thus disabling the HV output. 



--- Code: ---MEMORY ADDRESS OF EPROM: 0000~1FFF (PROGRAM SPACE:8KB,ST M2764A)

SRAM: 4000~47FF (DATA SPACE:2KB,HITACHI HM6116)

IRQ_ROUTIINE:  0038h
SOURCE:  SHEET 6
CLK->(U610)->TIMER @ 976.56hz/0.001024ms   (4Mhz/4096(U610)74HC4020)
GPIB

U201 -> 555 TIMER -> PIN 3
SWITCHING FREQ = 39.925Khz FOR HV TRANSFORMER


SRAM ADDRESSES:
SRAM_MISC = 413AH
SRAM_STATUS1 = NA
SRAM_STATUS2 = 401CH

EPROM ADDRESSES:
*IDN? STRING                 000019fah: 53 74 61 6E 66 6F 72 64 52 65 73 65 61 72 63 68 ; StanfordResearch
00001a0ah: 53 79 73 74 65 6D 73 2C 50 53                   ; Systems,PS

191CH GPIB OFFSETS TABLE 0000191ch: 24 35 19 2A 3B 19 48 6E 19 49 79 19 53 89 19 54 ; $5.*;.Hn.Iy.S‰.T
CMD CHAR+OFFSET 0000192ch: 8F 19 56 9A 19 57 AA 19 00                      ; .Vš.Wª..

193BH GPIB COMMAND 00001935h: 44 41 43 BE 1C 00 43 4C 53 6A 1A 45 53 45 3C 1E ; DAC¾..CLSj.ESE<.
CMD+OFFSET+00H 00001945h: 45 53 52 AE 1E 49 44 4E B0 19 50 53 43 72 1E 52 ; ESR®.IDN°.PSCr.R
00001955h: 43 4C 5A 1D 52 53 54 4E 1A 53 41 56 35 1D 53 52 ; CLZ.RSTN.SAV5.SR
00001965h: 45 06 1E 53 54 42 08 1F 00 56 4F 46 14 1A 56 4F ; E..STB...VOF..VO
00001975h: 4E 24 1A 00 4C 49 4D 7F 1B 4F 55 54 7E 1C 54 52 ; N$..LIM.OUT~.TR
00001985h: 50 EE 1B 00 4D 4F 44 9E 1C 00 43 4C 52 3B 1A 4D ; Pî..MODž..CLR;.M
00001995h: 4F 44 FF 1C 00 4C 49 4D 06 1B 4F 55 54 5D 1C 53 ; ODÿ..LIM..OUT].S
000019a5h: 45 54 87 1A 00 4F 52 44                         ; ET‡..ORD


19FAH 000019fah: F3 CD D5 11 FB 0E 00 C3 D4 15 CD 9D 17 C2 FD 17 ; óÍÕ.û..ÃÔ.͝.Âý.
                                                        00001a0ah: 21 8D 1F 11 3C 41 01 09 00 ED B0                ; !..<A...í°     

1F7BH 00001f7bh: 5A 31 33 34                                     ; Z134
1F7FH EPROM_VERSION_NUMBER 00001f7fh: 31 2E 33 34                                     ; 1.34

1F83H                                               00001f83h: 88 13 88 13 82 14 82 14 64 00                   ; ˆ.ˆ.‚.‚.d.
1F8DH 00001f8dh: 00 00 88 13 82 14 82 14 00 00 00 00 00 00 00 00 ; ..ˆ.‚.‚.........
00001f9dh: 0E 48 0D                                        ; .H.

1FA0H EPROM_CAL_DATA 00001fa0h: AC 09 E4 55 0B 52 C4 09 54 52 54 52 BC 09 19 52 ; ¬.äU.RÄ.TRTR¼..R
00001fb0h: 53 4E BB 09 28 52 55 4E AC 09 E6 BF E2 C1 BD 09 ; SN».(RUN¬.æ¿âÁ½.
00001fc0h: E0 C1 E4 C3 00 00 93 43 93 43 A6 09 96 55 F6 51 ; àÁäÃ..“C“C¦.–UöQ
00001fd0h: C4 09 54 52 54 52 D1 09 31 51 30 4D D0 09 52 51 ; Ä.TRTRÑ.1Q0MÐ.RQ
00001fe0h: 22 4D A6 09 12 C0 E2 C1 D2 09 78 C2 84 C4 00 00 ; "M¦..ÀâÁÒ.x„Ä..
00001ff0h: 93 43 93 43 00 00 76 CA 00 00 76 CA             ; “C“C..vÊ..vÊ

1FFCH EPROM_SERIAL_NUMBER    00001ffch: 31 31 33 37                                     ; 1137


GPIB
DAC  CLS/*CLS ESE/*ESE  ESR/*ESR
IDN/*IDN  PSC/*PSC RCLZ/*RCL RSTN/*RST
SAV/*SAV  SRE/*SRE STB/*STB  VOF/HVOF
VON/HVON  LIM/ILIM OUT/IOUT  TRP/ITRP
MOD/SMOD  CLR/TCLR MOD/TMOD  LIM/VLIM
OUT/VOUT  SET/VSET ORD


I/O ADDRESSES:

U504  (A3-A5 -> A-C, A6-G1, WR->!G2A, IORQ->!G2B )
40H: SPARE
48H:-SETS -> U407/U408
------------------------------
BIT0: U407A / GND->U405.3
BIT1: U407B / U410.RF -> U405.3
BIT2: U408A / ANALOG_VOLTAGE
BIT3: U408B / I_OUT_MONITOR
BIT4: U408C / CV_SET
BIT5: U408D / DV_LIM
BIT6: U407C / I_LIM
BIT7: U407D / I_TRIP
------------------------------

50H:-DAC U410 12BIT DAC AD7542
50H  DAC_LOW_NIBBLE
51H  DAC_MID_NIBBLE
52H  DAC_HIGH_NIBBLE
53H  DAC_LOAD

58H:-LED U611 74HC374
------------------------------
BIT0: L0
BIT1: L1
BIT2: L2
BIT3: -AD1/V_MON 
BIT4: -AD2/I_MON
BIT5: -AD3/ANALOG_VOLTAGE_IN
BIT6: -IO
BIT7: UNUSED
------------------------------

60H:-SEGB U606 74HC374
------------------------------
BIT0: EDP
BIT1: EG
BIT2: EF
BIT3: EE
BIT4: ED
BIT5: EC
BIT6: EB
BIT7: EA
------------------------------

68H:-SEGA U609 74HC374
------------------------------
BIT0: ODP
BIT1: OG
BIT2: OF
BIT3: OE
BIT4: OD
BIT5: OC
BIT6: OB
BIT7: OA
------------------------------

70H:-STROBE     (STROB0-STROB5 LED)  (STROB2-4 KEYBOARD SCAN OUT)
------------------------------
BIT0: STROB0  (LED)
BIT1: STROB1 (LED)
BIT2: STROB2  (LED,KBD)
BIT3: STROB3  (LED,KBD)
BIT4: STROB4  (LED,KBD)
BIT5: STROB5  (LED)
BIT6: STROB6  ()
BIT7: STROB7  ()
------------------------------

78H: MISC -> (U507,74HC374 OCTAL D-TYPE FLIP-FLOP) >CLK
------------------------------
BIT0: -SHUTDOWN
BIT1: UPOK
BIT2: -FLAG_RESET
BIT3: -TIMER_RESET
BIT4: (SPARE)
BIT5: -POS
BIT6: -NEG
BIT7: FILTER
------------------------------

U505 (A3-A5 -> A-C, !M1->G1, A6->!G2A, IORQ->!G2B )
00H: SPARE
08H: SPARE
10H: SPARE
18H: SPARE

20H: STATUS1 U506->74HC244
------------------------------
BIT0: COMPARE
BIT1: I_FAULT
BIT2: CUR_LIM
BIT3: PRI_FAULT
BIT4: V_FAULT
BIT5: TIMER
BIT6: -J504/SET_MONITOR_SW
BIT7: GND
------------------------------

28H: STATUS2 U512->74HC244
------------------------------
BIT0: HV_ON
BIT1: HV_OFF
BIT2: J501/-HV_SW_POS
BIT3: J502/-HV_SW_NEG
BIT4: GND
BIT5: GND
BIT6: GND
BIT7: GND
------------------------------

30H: -KBD
------------------------------
BIT0: KBD0
BIT1: KBD1
BIT2: KBD2
BIT3: KBD3
BIT4: KBD4
BIT5: KBD5
BIT6: KBD6
BIT7: KBD7

key press display
01 Man         
02 GPIB       
03 left-arrow 
04 Right arrow
05 up-arrow   
06 down-arrow 
07 Enter       
08 Select     
09 STO         
10 RCL         
11 CLR         
12 .           
13 0           
14 7           
15 4           
16 1           
17 8           
18 5           
19 2           
20 9           
21 6           
22 3           
------------------------------

38H: -GPIB
------------------------------
READ
38H GPIB_INT_STATUS_0
BIT0: MAC  (ADDRESS CHANGE)
BIT1: RCL  (REMOTE/LOCAL CHANGE)
BIT2: SPAS (RSV1/2 SERVICE REQUEST)
BIT3: END  (LAST BYTE RECEIVED)
BIT4: BO    (BYTE OUT)
BIT5: BI    (BYTE IN)
BIT6: INT1 (1 WHEN BITS 2-7 OF STATUS 0 SET)
BIT7: INT0 (1 WHEN STATUS 1 BIT SET)
39H GPIB_INT_STATUS_1
BIT0: IFC (INTERFACE CLEAR)
BIT1: SRQ   (SERVICE REQUEST)
BIT2: MA    (MY ADDRESS)
BIT3: DCAS  (DEVICE CLEAR ACTIVE STATE)
BIT4: APT  (ADDRESS PASS THROUGH)
BIT5: UNC   (UNRECOGNIZED COMMAND)
BIT6: ERR (ERROR)
BIT7: GET (GROUP EXECUTE TRIGGER)
3AH GPIB_ADDRESS_STATUS
BIT0: ULPA
BIT1: TADS
BIT2: LADS
BIT3: TPAS
BIT4: LPAS
BIT5: ATN
BIT6: LLO
BIT7: REM
3BH GPIB_BUS_STATUS
3EH GPIB_CMD_PASS_THRU
3FH GPIB_DATA_IN

WRITE
38H GPIB_INT_MASK_0
39H GPIB_INT_MASK_1
3BH GPIB_AUXILIARY_CMD
3CH GPIB_ADDRESS
3DH GPIB_SERIAL_POLL
3EH GPIB_PARALLEL_POLL
3FH GPIB_DATA_OUT
------------------------------

--- End code ---

Navigation

[0] Message Index

[#] Next page

[*] Previous page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod