So it turns out the display is just fine, all segments operating normally.
I assumed the internal calibration factors would have been lost due to the age of the 3v lithium batt, but they are intact. The batt looks sort of like a cr123, I wonder if I could simply replace the original with a fresh cr123? Amazingly it still shows 3v on my DMM.
I bought this because it was cheap and in very nice cosmetic condition, but also to use as an additional comparison for my Fluke 332D. With the 332D set to 1v the Dana 5004 dithers 20-40 counts above and below 1.00000 but it never settles. I am wondering if this could be a fault with the meter. I have been browsing the manual and it appears it should be able to display a stable reading.
If this is indeed a problem, does anyone have advice on where to look first?
I did notice three large electrolytic caps (C82, C80, and C77) on the mother board show evidence of heating, their outer plastic shrink wrap coating has melted through in spots. I checked them in circuit with my component tester and they test in spec with low ESR. So? I think these are PSU filter caps so could ripple impact the stability of the meter?
Edit: I checked the output of the 332D on my scope I am seeing around 16-20mv p-p ripple.