Success!
Checked all replaced components for orientation / type 3x. Partially re-assembled K2002, pushed power button on.
Meter springs to life. Added bonus: all self-tests pass. Seems like no errors. I let it run ~10 hours over night, re-ran self tests again and all self-tests pass. Preliminary check done on 10V range:
The 34401A in the photo was last calibrated 6/2014. Ignore the K2700 above ... seems out of cal ...
Well I like the fact that the K2002 I restored ( last cal: 2004! ) matches my most recently cal'd 34401A, at least to 6.5 digits. Especially since I removed the thermal sponge and then cleaned with IPA the area left with all the goo from the sponge/adhesive.
Some observations:
-- U105 (5V reg) gets uncomfortably hot to the touch. I can see now why the cap next to it tends to leak/explode. I am going to put a low profile heat sink on U105, it is in the air channel anyway.
-- Testing 200mV range: shorted HI to LO with bare copper wire, pushed into contact. Tried all combinations of filtering, A/D sampling window, Line Sync etc but meter seems very noisy. Readings jumping around between 100uV - 800uV. I have the metal case cover off and I didn't put the ferrite beads back on the input wires yet ... not sure if that would make this much noise. I can post a video soon of this behavior soon.