Author Topic: Repair Fluke 5500A Error 104  (Read 333 times)

0 Members and 1 Guest are viewing this topic.

Offline Regiano

  • Newbie
  • Posts: 2
  • Country: br
Repair Fluke 5500A Error 104
« on: April 13, 2019, 04:32:49 pm »
I had a problem and entered 220V at the input of my fluke 5500A Calibrator , now it is turning on ERROR 104 trip relay, can someone help me with this defect.
thanks
 

Offline dacman

  • Frequent Contributor
  • **
  • Posts: 396
  • Country: us
Re: Repair Fluke 5500A Error 104
« Reply #1 on: April 15, 2019, 02:31:05 am »
I would suggest sending it to Fluke for repair.  I know someone that zapped a 5520A and Fluke replaced a plethora of items in it.  (The 5502A and 5522A have input protection.  The 5500A and 5520A do not.)
 
The following users thanked this post: Regiano

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 2064
  • Country: ca
Re: Repair Fluke 5500A Error 104
« Reply #2 on: April 15, 2019, 03:01:07 am »
services manuals with schematics,   surely you have them already  ??

https://archive.org/details/FLUKE5500ASERVICEMANUALWITHSCHEMATICS/page/n269

you can choose different format,  the pdf is around 12 megs

https://www.dropbox.com/s/48p5qlqob1f4ysj/FLUKE%205500A%20SERVICE%20MANUAL%20WITH%20SCHEMATICS.pdf?dl=0

the input sections are at section 6-12  page 263 ...

do you have something burned  busted  cracked  to help pinpoint a section ??
« Last Edit: April 15, 2019, 03:11:18 am by coromonadalix »
 
The following users thanked this post: Regiano