For the J202 connector, I know the details for the base pin now. Because I opened another F5522A.
But I met a problem for this one, it showed "Error 1506: IDAC counts out of range" when set to OHM simulation
, and showed "Error 399: Fault during DVZ" during the calibration by itself
Does anyone have any suggestions and resolved the same issue or have any service info?
I checked the CR40 and CR41 (Type: MBRS140T3G,Vendor: On semi, Marking code: B14, a 40V 1A Schottky Diode) were damaged with A5 impedance synthesizer, whatever the positive or negative direction, the voltage drop measured result was 0.000V, so the diodes were short circuit.
I have to this point replaced Vishay's SS14 for CR40/CR41 as above., but the Error is still there.
Thanks a lot.
With best regards.
From Seebeck.