In this thread I like to show you a huge problem with the R&S RTB2004 and probably some other scopes and how to repair it.
But history first!
Some years ago I bought a new and shiny RTB2004. After two years I noticed that I had a huge offset drift in one or two channel. I don't
understand the reason, but it was just in warranty, so I send it back and R&S repaired it. Repairing means, they exchange
the mainboard and calibrate it. So you got a more or less new scope.
Another two year later I found the offsetproblem again. But this time I understand the reason. I used a selfmade current probe to
measure current of my DUT. In most cases I use this probes on channel 3 or 4, and more seldom on CH1.
So my current measurement destroyed the offset. But why?
My probe has a OPV at the output (Vcc=3.3V) and delivers 0-2.5V for any current range. I am measuring the current consumption
of some microcontroller board. So I put the GND line at the bottom of the screen and the 2.5V at the top.
Most of the time the scope is now measuring 2V on the top of the screen in DC mode.
The internal PGA inside the scope a LMH6518 has to withstand the offset for some time, but it can't.
TI find it is to much for the PGA. They changed the datasheet later. They did not alow a huge DC offset
anymore. Unfortunatly the developer at R&S did not know it.
I had a diskussion with TI about it:
https://e2e.ti.com/support/amplifiers-group/amplifiers/f/amplifiers-forum/1569255/lmh6518-offset-problem-with-lmh6518However there are some more discussion about this problem, because I am not the first victim.
This time R&S did not repair it in warranty because they only give 1 year of warranty for there repairs.
So I decide it to repair it by myself! That means replacing the LMH6418
But before, have a look of some picture I did BEFORE repairing. It is important to memoryze it to understand what happens at the end.
The picture show you the huge offset drift before repairing and the that the internal offset compensation stops with an error
after 25%. That is important!
Olaf