Author Topic: repair of the intensity control on a tektrionx 317 scope  (Read 962 times)

0 Members and 1 Guest are viewing this topic.

Offline KV8ETopic starter

  • Newbie
  • Posts: 4
  • Country: us
repair of the intensity control on a tektrionx 317 scope
« on: April 29, 2018, 08:35:00 pm »
I have a Tektronix 317 scope. the scope works fine except the intensity control has no effect on trace brightness. I have checked the voltages and the 400v unregulated supply is running at 450-460v. the other supply voltages r very close to correct. the voltage across the brightness control pot goes from +40v to 0v with the pot at the most cw position. seems to me the voltage must go - to dim the trace. other than this the scope works fine. any help on this problem??
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 709
  • Country: us
 

Offline KV8ETopic starter

  • Newbie
  • Posts: 4
  • Country: us
Re: repair of the intensity control on a tektrionx 317 scope
« Reply #2 on: May 06, 2018, 08:08:52 pm »
think I have and answer to my intensity control problem. the rectifier tube v824 may be bad. I ordered replacements and will put it in when they arrive. will keep all informed of process.  kv8e   doug
 