Hi all,
I picked up one of these fairly cheap as a defective unit. Opening the unit revealed two things: 1) Old, leaking electrolytic caps, and 2) flickering neon bulbs in the chopper circuit.
I managed to replace the neons with NE2U tubes, no fancy aging or testing for ignition voltage etc (though I may need to revisit this). I also replaced the worst offending caps (10uf, originals were long gone and leaking electrolyte), I am waiting on the remainder from mouser.
My unit was stable for a while and then began to peg in the 3uv and 1uv range in operate mode, in zero mode it would still zero. I opened the unit to discover flickering neons (the new ones
). But, the reason for my pegging turned out to be the operate/zero switch. Cleaning the operate contacts allowed the unit once again achieve zero in operate mode in the 10, 3 and 1uv ranges. I imagine the range switch contacts are another source of potential problems...
My question now is, do these neon based choppers tolerate some flicker as part of the design? Has anyone opened their functional unit to discover one of the neons flickering?