Bonjour, Normally A5 SMD cap leakage (there are may of the caps) damages nearby components and eats internal traces on the PCB. The fluid causes the copper to becone unslderable, give away i a odd smell on heated traces and joints.
Suggest a very deep inspection, removal all affected parts. Total clean of board, check trace continuity.
If the DAC and REF ckt is not perfect, all sorts of control oddities, as the scope is "fly by wire"
See very long threads on this in the old 2465B teardown, also excellent Groups.io TEKSCOPES, and TEKSCOPES2
ask there, many ex TEK experts.
https://groups.io/g/TekScopes/topicshttps://groups.io/g/TekScopes2/topicsBon Chance,
Jon
PS: A5 had the notorious Dallas 1225 NVRAM with dying /dead Li batter that holds all CAL.