Please allow me to ask, I'm having a problem, suddenly my Tektronix TDS220 oscilloscope doesn't produce an image or a white screen, the CFL light is on. I tried checking the PIN 1 voltage (Vo = LCD Contrast adjust voltage) on the LCD connector which is equal to 0 volts, and the PIN2 voltage (Vee = Power supply for LCD (-24V)) is also 0 volts, please give me instructions and guidance on what other steps to take. what should I fix?
I really hope to be able to revive the oscilloscope.
Thank you for all the instructions and guidance
 

Offline m k

  • Super Contributor
  • ***
  • Posts: 2038
  • Country: fi
Re: Repair tektronix TDS 220 no screen (white screen)
« Reply #1 on: April 21, 2024, 02:34:33 pm »
You should dig up the missing voltage.

If no schematics is available you must trace the route back the hard way.
Advance-Aneng-Appa-AVO-Beckman-Data Tech-Fluke-General Radio-H. W. Sullivan-Heathkit-HP-Kaise-Kyoritsu-Leeds & Northrup-Mastech-REO-Simpson-Sinclair-Tektronix-Tokyo Rikosha-Triplett-YFE
(plus lesser brands from the work shop of the world)
 

Offline rh100605

  • Regular Contributor
  • *
  • Posts: 56
  • Country: gb
Re: Repair tektronix TDS 220 no screen (white screen)
« Reply #2 on: April 26, 2024, 01:48:31 pm »
Hi,
Did you measure all the power supply voltages at the socket on the main board. The specifications for each pin are in the service manual.
If they are not correct then unplug the power connector and measure the outputs of the power supply again.


 