Trigger
source ch 1 line ac couple
This would imply you've set the triggering to
Line, this means the scope will trigger at your mains line frequency. This is fine for when needed, but that is not often. Turn Line triggering off.
For most use you'll want the trigger DC coupled, change that too.
X5 s/div multiplier should be off too.
With a CRO you need think about the signals you are intending to see, they're not at all like a DSO where you can just hit the Autoset button and it sorts it all out for you. You do have to "drive" a CRO.
First sanity check is the probe Cal output, they're mostly 1 KHz frequency and square wave and at a few volts, commonly 3 or 5V. So now set the vertical to 1V/div and the timebase to 1ms/div and your ready to check the probe Cal output with a 1:1 probe. Remember a 10x (10:1) probe will divide the signal by 10 so to see the same amplitude waveform as you've just set up for, the vertical would need to be changed to 100mV/div.
Got your head around all that?
So have a play with a piece of wire in the BNC and your probe Cal output, but nothing else until your probes arrive.

Again, think about the settings you'll make before anything's connected, probably Trigger settings are the most important.
Take the time to watch some of Alan's vids on scope use in the sticky thread in the Beginners forum.
Enjoy.