I obtained a broken Fluke 845AR, with wild meter swings. Upon opening and inspecting the neon chopper bulbs, they were a little dim, and blackened.
I replaced them with some NE2U A3C tubes that were readily available locally. This alone has brought the meter back to life and now I can zero it, with around +- 50 nanovolt noise swings on the most sensistive 1 microvolt range. Is this about normal?

Some further notes:
I pulled the old tubes out and made a comparison of their behavior compared to the old ones. With a 30K resistor in series, and the room darkened, the new ones ignite around 12 volts lower (104v vs 116v) than the old ones and of course have clear glass. I also noted that originals lit up the entire electrode inside, whereas the new ones light only in the center. Therefore, I measured a small amount of extra height was needed to align this Illumination with the center of the glass tubes. Pictures of old and new attached.