Author Topic: Repairing Tektronix 2213  (Read 1857 times)
Offline gerovite
  • Country: by
« on: October 20, 2015, 01:12:51 pm »
I appeal to owners Tektronix 2213
What should be the waveform in point 9 SWEEP GENERATOR & LOGIC
thanks in advance
Offline singapol
  • Country: sg
Re: Repairing Tektronix 2213
« Reply #1 on: October 21, 2015, 08:44:16 am »
There is nothing wrong with the waveform except that maybe your oscilloscope's rise time/bandwitdh is not fast enough. Try using X10 setting on probe. The 74Sxx series logic is pretty fast.
Offline oldway
  • Banned!
« Reply #2 on: October 21, 2015, 08:49:11 am »
Welcom on the forum.
It would be better to describe the problem you have with your scope.
What's wrong ?
Offline gerovite
  • Country: by
« Reply #3 on: October 21, 2015, 09:54:13 am »
Thanks for your kind reply :)
I can not take a picture on the oscilloscope because currently there is no chip 74f74.
 Here is the problem:
  When capture sin in CRT displayed:
Offline oldway
  • Banned!
« Reply #4 on: October 21, 2015, 10:04:53 am »
That's not a very clear explanation...Is the sine wave sincronised ?
Do you display only one channel or both ?
If sine wave is well syncronised and only one channel displayed, it seems to be a blanking problem during retrace.
Offline gerovite
  • Country: by
« Reply #5 on: October 21, 2015, 10:13:16 am »
sorry, it's only one ch
Offline oldway
  • Banned!
« Reply #6 on: October 21, 2015, 10:27:19 am »
Trigger works well or not ?
Offline gerovite
  • Country: by
« Reply #7 on: October 21, 2015, 10:32:02 am »
Trigger work
« Last Edit: October 21, 2015, 10:53:59 am by gerovite »
Offline oldway
  • Banned!
« Reply #8 on: October 21, 2015, 10:45:58 am »
In repair, there is a basic rule: always to start from the beginning and never to rely on the diagnostic made by others.
What You suspect does not interest me, what I want is a detailed description of the problem as it appeared originally.
I do not need you to copy the service manual, I have it too.
Simply answer to my questions !
With your way of responding, whe are going nowhere. :palm: