Electronics => Repair => Topic started by: Mani556 on January 02, 2025, 08:39:36 am
Title: Replacement of a Chopper Transformer
Post by: Mani556 on January 02, 2025, 08:39:36 am
i am biomedical engineer in a hospital and i am repairing an EMG Machine, firstly it was not generation current for stimulation of muscles i opened the board and found that 1 chopper transforemer is faulty.
written on it is Cramer 1826 A2273.
i have checked online and also on the cramer website but i cannot find replacement for this chopper.
can anyone help me guide me in this regard please.
i shall be very thankful.
Title: Re: Replacement of a Chopper Transformer
Post by: squadchannel on January 02, 2025, 08:53:57 am
:wtf: Do not tamper with medical equipment. |O
Post by: Xena E on January 02, 2025, 09:40:13 am
https://cramercoil.com/
Use the enquiry form, if they cant help, no one will.
Post by: Mani556 on January 02, 2025, 09:43:24 am
i already did, but no reply
Post by: Xena E on January 02, 2025, 11:07:22 am
If this item has value, consider getting it repaired/rewound.
Costly though.
Crammer will reply to most enquiries, eventually. In this instance though they will probably direct you to the manufacturer of the device if it was a bespoke design.
X
Post by: inse on January 02, 2025, 11:20:00 am
How did you determine the transformer being at fault?
You may post a photo of the device as well.
Post by: Mani556 on January 02, 2025, 11:40:10 am
it is not giving continuity on one side
Post by: inse on January 02, 2025, 11:47:05 am
Probably the winding is broken at the pin, maybe it can be fixed if accessible