i am biomedical engineer in a hospital and i am repairing an EMG Machine, firstly it was not generation current for stimulation of muscles i opened the board and found that 1 chopper transforemer is faulty.
written on it is Cramer 1826 A2273.
i have checked online and also on the cramer website but i cannot find replacement for this chopper.
can anyone help me guide me in this regard please.
i shall be very thankful.
Online squadchannel
  • Country: jp
  • deepl translate user
Re: Replacement of a Chopper Transformer
« Reply #1 on: January 02, 2025, 08:53:57 am »
:wtf: Do not tamper with medical equipment. |O
Offline Xena E
« Reply #2 on: January 02, 2025, 09:40:13 am »
https://cramercoil.com/
Use the enquiry form, if they cant help, no one will.
  • Country: ae
« Reply #3 on: January 02, 2025, 09:43:24 am »
i already did, but no reply
Offline Xena E
« Reply #4 on: January 02, 2025, 11:07:22 am »
If this item has value, consider getting it repaired/rewound.
Costly though.
Crammer will reply to most enquiries, eventually. In this instance though they will probably direct you to the manufacturer of the device if it was a bespoke design.
X
Offline inse
« Reply #5 on: January 02, 2025, 11:20:00 am »
How did you determine the transformer being at fault?
You may post a photo of the device as well.
  • Country: ae
« Reply #6 on: January 02, 2025, 11:40:10 am »
it is not giving continuity on one side
Offline inse
« Reply #7 on: January 02, 2025, 11:47:05 am »
Probably the winding is broken at the pin, maybe it can be fixed if accessible