Hello everyone, first post...
I am trying to fix a burnt out output attenuator in an Agilent 33120A. The resistors I need to replace are 69 ohm, 0.1%, 1206's, and these seem almost impossible to get a hold of in small quantities. Anyone have any good ideas as to where I could get hold of just a couple of these??
Offline KJDS
    • my website holding page
Re: Resistors for 33120A
« Reply #1 on: November 18, 2014, 08:35:54 pm »
Have you tried some calculations of the affect of using a readily available resistor?
  • Country: no
« Reply #2 on: November 18, 2014, 08:53:03 pm »
As far as I understand it would directly affect the accuracy of the output amplitude from the function generator, 1Vpp on the display would no longer equal 1Vpp output. But I haven't done any calculations as to the degree of change with other values, just wanted to see if it is at all possible to get a hold of the correct values first...
« Reply #3 on: November 18, 2014, 09:09:20 pm »
Why not stack 3 x 210 Ohm 1206 ?
Get say 10 x and select for value to get a spot on combination.
Offline brillopad
Offline lowimpedance
  • Watts in an ohm?
« Reply #5 on: November 18, 2014, 10:16:34 pm »
Stacking is the best way to get what value you need, but be mindful of the temp co. as well.
  • Country: no
« Reply #6 on: November 19, 2014, 10:48:47 am »
Ok, so I have decided to stack one 69.8 and one 6.04k, should give 69.003 total and both of these values are pretty easy to get hold of. My plan was getting .1%, 25ppm resistors, those specs should match the manufactureres parts acording to the manual.
One thing, is thin film the best option since we are dealing with freq up to 15MHz??
  • Country: no
« Reply #7 on: November 28, 2014, 07:17:41 pm »
OK, new resistor in place and it works great, didn't even have to recalibrate it.
I also replaced the U205 SRAM to fix a 604 error, and that kind of worked. The 604 error disappeared, but now I get a 606 and 607 error. (They were probably there all along, just the self test that stopped at the 604 error.) Anyone have any idea where to start troubleshooting this??