0 Members and 1 Guest are viewing this topic.
I’ve been restoring a Voltohmyst along with several other VTVM’s and tube testers and a couple of vintage but solid state multimeters. My advice with any equipment repair or restoration of devices with a moving coil meter that is decades old is to first, before anything else to test the meter movement itself to see if it can still reach full-scale-deflection at the stated FSD current. It’s not that uncommon for some old meter movements to lose field strength in their permanent magnet and in some test gear this makes it unable to be calibrated as the meter just can’t deflect enough. Be very careful though never to try and test a meter movement with a multimeter ohms range, many movements have FSD current in the uA range so you need a power supply and a decade resistance box and a precision resistor to measure current drop and a good multimeter to measure that drop and to calculate the meter coil resistance. Putting a multimeter on ohms range directly across the meter will damage it as the multimeter test voltage is too high.