Hi dah5477, thanks tremendously for sending me this information.
I confirmed with a coworker the board appears repairable. The resistor was laser trimmed, and there are three pads internally. We attempted to measure the resistance of the burned resistor but wasn't able to get a reading less than 7k (obviously wrong). After scratching away some of the material near the edges we could measure the resistance of each internal pad to be 38.6 ohms, 34.7 ohms, and 60.7 ohms, without taking into account the resistance of the probes. This placed the parallel resistance at a minimum of 14 or 13.7 ohms. I saw the other resistor on the board was a 12 ohm resistor and so suspected we were looking at that value in reality. Looking at your board I can tell our strategy got us at least in the ball park.
Googling H2 markings located the GaAs transistor to be SHF-0289. It originally was made by a company called RFMD but is now Qorvo. It's obsolete, with some venders indicating stock. Last thing I did this evening was measure the continuity of these leads, and sure enough one of the components has the gate shorted to the drain / source. The other GaAs transistor only has the drain and source shorted together which makes sense as this is a depletion part from the datasheet. It's interesting your markings indicate FP2189 - can you give us a picture of the other side of your board? I recognized in my search earlier an alternative marking to H2 was to list the part number and some other digits.