Title: Rigol DS1054z offset problem
Post by: martin.s on March 10, 2016, 06:21:36 pm
Hello there,

I have a problem with my Rigol DS1054Z.

When I measure DC signal with CH2 I found problem with offset.
You can see it in print-screen and photos.

I tried to change probes, re-calibration in setup menu.. everything else.. and phenomenon still persist.  :-\

Without any signal there is without any offset.

Do you have some solution for it?
Title: Re: Rigol DS1054z offset problem
Post by: Bud on March 11, 2016, 03:25:45 am
You may get a clue from Part 4 of my thread here once I post it:

https://www.eevblog.com/forum/projects/project-yaigol-fixing-rigol-scope-design-problems/ (https://www.eevblog.com/forum/projects/project-yaigol-fixing-rigol-scope-design-problems/)

But you have to wait a bit for it.
Title: Re: Rigol DS1054z offset problem
Post by: JFJ on March 11, 2016, 01:16:48 pm
I tried to change probes, re-calibration in setup menu...
Have you tried re-calibrating with the channel inputs shorted to ground? Also, don't forget to let the oscilloscope warm-up for 30-minutes before re-calibrating.
Title: Re: Rigol DS1054z offset problem
Post by: Mechatrommer on March 12, 2016, 04:57:40 am
overdriven input..
https://www.eevblog.com/forum/testgear/mechatrommer's-rigol-bug-report-space/msg883553/#msg883553 (https://www.eevblog.com/forum/testgear/mechatrommer's-rigol-bug-report-space/msg883553/#msg883553)
http://cp.literature.agilent.com/litweb/pdf/5989-0068EN.pdf (http://cp.literature.agilent.com/litweb/pdf/5989-0068EN.pdf)