I guess I should say that I also have never done signature analysis before, or used any logic analyzer. Just the schematic, maybe a pinout/data sheet and a 'scope.
anyway, the real fun begins, exploring the circuits, and the idea of having a new tool to assist in the process is refreshing.
So we begin at the Basic Processor test loop - Just the A7, A9 and A10 boards in the card nest. Follow the setup for the SA and probe +5v. and get 9PA2
Great! Now we check the IDA Bus signatures, and they match exactly the table provided in the service manual.
Now move to the Primitive ROM test.
The manual states that to enter the primitive ROM test, install A6 board (ROM), move jumper A7 (J2) to run and power on, then short out A7(J4) and press RESET on the front panel - Keep A4 shorted. +5v signature = OH62 should appear in about 10 seconds.
My board did not ever give a signature. But that is OK, because there is a pretty extensive ROM SA listing that covers address, chip select, and Data lines.