Hi,
I've got a CMU200 that becomes non-responsive when the RF generator option is turned ON. I've dig into the problem and so far I found the following. I have replaced the original HDD with a SSD by copying it with HDDGuru Raw copy tool and that seems to work and CMU boots normally.
1. The CMU200 takes a long time to load the options and stays at BaseDiscoveryOptionsDevice. This is true for both hard drives and it makes no difference in drive being HDD or SSD.
a. I have noticed that on another blog thread this has been noticed. The link is below.
https://www.eevblog.com/forum/testgear/rohde-schwarz-cmu200/150/ The mentioned CMU200 in the above thread had a broken DIG module (TXDSP MODULE in R&S jargon) and replacing that corrected the RF generator freeze.
2. System self test on RF front end shows some voltages out of range. Check picture-1
3. System self testing RXTX board shows the same voltages referred above are out of range Check picture-2
4. System self test shows that the CMU-B68 board is not usable. I think this is the main culprit here. Check picture-3
I've already taken out the CMU-68 and CMU-B21 combined board and reseated it, in case to see if that might be the problem. But no change. The board is still not usable according to the self test.
5. The CMU200 becomes non-responsive when the system self test reaches the CMU-B56Var14 test. Check picture-4
6. FE self test shows over voltages. Check picture-5
7. REF self test is okay. Check picture-6
8. DIG self test is okay. Check picture-7
9. RXTX self test shows over voltages. Check picture-8
10. Hardware data verification test does not show passed for the CMU-B68. Check picture-9
11. Options at the start up shows the CMU-B68 is available. Check picture-10
12. Initializing the WCDMA FDD option shows a software error and it says one of the C++ files are missing. Check picture-11
13. L1CoPro Level calibration makes the CMU to freeze and become non-responsive.
Please support to solve this problem.
Thank you in advance.