Author Topic: SDG1025 Damaged CH1 Repair  (Read 444 times)
Offline hesam.moshiriTopic starter
    • MyVanitar
« on: November 18, 2024, 05:12:23 pm »
I'm trying to repair my Siglent SDG1025 function generator which has damaged CH1. CH2 works properly.
I replaced two components (U19 and Q3=BAV99) but channel one shows a weird waveform.
U19 = THS3095 is a special OpAmp that can be turned ON-OFF via its PIN-8. so it's likely that it's OFF.
I spent a lot of time, but I could not find a trace connecting to PIN-8 of U19. I can see a tiny track and a VIA on the PCB connected to PIN-8, but it goes from mid-layers to somewhere unknown.
if somebody has this device please check the voltage on PIN-8 of U19, when channel-1 is active and Vpp is above 2 volts.
Offline fzabkar
Re: SDG1025 Damaged CH1 Repair
« Reply #1 on: November 18, 2024, 08:07:28 pm »
https://www.ti.com/lit/ds/symlink/ths3091.pdf (datasheet)
As I see it, you need to compare the PD* voltage on pin #8 against the REF voltage on pin #1.
    • MyVanitar
« Reply #2 on: November 18, 2024, 08:40:24 pm »
REF (Pin-1) is grounded, so I'm looking for the PIN-8 connection on the board. The voltage on Pin-8 is low (max 0.4V)
« Last Edit: November 18, 2024, 09:13:14 pm by hesam.moshiri »
Offline fzabkar
« Reply #3 on: November 18, 2024, 11:54:44 pm »
Are you able to trace pin-8 of the working channel back to its source?
    • MyVanitar
« Reply #4 on: November 19, 2024, 07:37:51 am »
for sure if it was designed like that, I was spotting it. This opamp is only used for CH-1
« Reply #5 on: November 21, 2024, 11:04:43 pm »