I was able to get my hands on an old Keithley 2001 DMM which failed calibration and could not be brought back into spec by full adjustment.
All electrolytic caps in the unit were replaced, but that did not help.
The following steps of the self-test fail:
- 200.1 A/D "No A/D comm, noisy A/D"
- 200.6 "A/D U808 malfunctioning"
- 200.7 "A/D U808 malfunctioning"
- 201.2 Testcal "Cannot measure 7V at A/D"
The meter displays reasonable values in all ranges, but the noise is higher than I expected, and there seem some strange non-linearities.
I connected the 2001 to 10 V from a calibrator, parallel to a Keithley 2010 which should be OK, took 500 readings at 10 PLC, and got the attached curves.
I also went through the voltage range from 0 V to 5 V using a DAC (not the calibrator) and recorded the deviation from the set value. The general curve shape is the fault of my DAC, but you can see the noise level of my K2001.
I have already built an adapter cable to get the ADC board out of the way, and probed around a bit, but did not find anything that looked obviously wrong to me. I don't have a deep understanding of what is happening inside the meter, and what I should be able to measure when single stepping through the self-test steps. The unit also does not seem to provide any more details about what it thinks is wrong...
Does anyone here have a suggestion on how to get this poor thing fixed?