Author Topic: Frequency counter Philips pm 6662  (Read 1719 times)
Offline SatanDream
« on: December 15, 2016, 08:45:33 am »
Hi, all
Unfortunately I made a false manipulation with an hf transceiver
Two hs components.
I seek the schema or the reference of the two components.
Thank you for Help
Offline yngndrw
Re: Frequency counter Philips pm 6662
« Reply #1 on: December 15, 2016, 01:14:35 pm »
Hi Rémy,
From a photo I have on my phone, it looks like the resistor is 68 Ohms and the capacitor next to it is 10nF - I'll confirm when I get home and open it up.
Thanks,
-Andrew.
Offline SatanDream
« Reply #2 on: December 15, 2016, 01:29:38 pm »
Hi, Andrew
 A big big thank you :scared: :clap:
Best regard
Online PA0PBZ
Offline SatanDream
Online PA0PBZ
« Reply #5 on: December 15, 2016, 04:04:14 pm »
Thank you but it is for pm6652 ;)
Oops...  :palm:
Offline yngndrw
« Reply #6 on: December 15, 2016, 08:01:27 pm »
Just opened it up to check, it's definitely a 68 Ohm resistor.
As for the capacitor, it says "10n K" on the top and has some text on the face towards the front of the counter, but I could read it without removing the whole board. It looks like the face with the text on it is the opposite side that burnt on yours so you should be able to read all of that. There's no text on the burnt side.
Hope that helps.
The following users thanked this post: SatanDream
Offline SatanDream
« Reply #7 on: December 16, 2016, 06:46:31 am »
Hope that helps.
Thank you for your kind assistance,
I was unable to find the service Manual.
For the resistance I will have to find the W, the capacitor normally 250v 10nf.
I will return to the forum to indicate the right values.
Best regard
« Last Edit: December 16, 2016, 06:52:34 am by SatanDream »
