Have both scope probes' compensation been adjusted for the respective channels?
A good starting point to understand which adjustments are which is to refer to the Tek 422 vertical preamp schematic and calibration procedure. [1] These classic analog scopes are all the same in this respect. Jumping into the middle of a calibration procedure can lead to undesirable results due to interaction of adjustments.
So, we shall start with the usual preliminaries.
First step is to adjust the step attenuator balance, for channel 1.
- Set sweep speed to 1ms/div.
- Select Auto trigger, set Trigger Level for free-running sweep.
- Set channel 1 input to ground, position trace on horizontal center line of display.
- Rotate the Volts/Div switch from 10mV to 20mV to 50mV and back. Ideally, the trace should not move up or down.
- R21 on the Tek 422 is adjusted to null the shift. On the M07192, the trimmer will likely be under the step attenuator switch.
* Tip: with the power off and before making any adjustments, measure and record the resistance from the wiper of each trimmer, to its ccw or cw end. This will facilitate resetting a control in the event it is not the intended location to be adjusted.
- Observe that adjusting the step attenuator balance trimmer causes a vertical shift of the trace.
- Re-center the trace with the vertical position control and again step from 10 to 50 mv/div and back, readjusting the balance trimmer as needed.
- Three or four repetitions may be required to reach the goal, due to interaction. The sensitivity of this adjustment can test one's patience.
- Once the balance is optimized, check the 5mV/div for minimal vertical shift.
Second step is to adjust the variable attenuator balance, for channel 1.
- Same preset conditions as above.
- Rotate the variable Volts/div control throughout its range and back. Ideally, there should be no vertical movement.
- R35 on the Tek 422 is adjusted to null the vertical shift. Again, on the M07192, the trimmer should be close to the back end of the Volts/div switch and connected to the variable attenuator potentiometer.
Repeat these two procedures for channel 2 adjustment.
A closeup side view photo of the step attenuators and variable controls can help identify the adjustment locations.
Before moving on to adjusting the gain and compensation of each channel, let's get these first two steps and controls sorted.
[1] Tek 422 service manual (135 MB) can be found here:
https://elektrotanya.com/tektronix_422_r422.pdf/download.htmlRefer to pages 85 & 86 paragraphs 8 & 9.