Author Topic: Siglent SDG1025 Self-Cal Failure  (Read 2371 times)
Offline sggtavTopic starter
  • Country: ar
« on: October 13, 2014, 10:07:14 pm »
Hi!
This is my first time posting on this forum!
I tried to run a self-adjust procedure on my Siglent SDG1025 Waveform Generator, and after it was done I found that the device stopped working correctly. When I turn on any of the 2 channels it looks as if they were "saturated", and if I change the voltage reference it modifies the level of that saturation but the signal never has the expected waveform.
And to make things even worse, I later found that my serial number was like "erased" during the process, because now it says "sdg000", while it should be like 14 digits long.
It seems as if I had lost all the configuration files or something like that. I wanted to know if this has ever happened to any of you, because I would really appreciate avoiding having to send it to Siglent.
I'll leave you a picture of the screen, where it shows the serial number problem.
I would really appreciate any piece of advice you can offer!
Thanks in advance!
Santi
Sggtav
Offline tautech
Re: Siglent SDG1025 Self-Cal Failure
« Reply #1 on: October 14, 2014, 06:02:17 am »
Santi, I will point Siglent China to you post.
Add your country flag to your profile.
« Last Edit: October 14, 2014, 06:05:16 am by tautech »
  • Country: ar
« Reply #2 on: October 14, 2014, 12:43:27 pm »
Thank you tautech, i appreciate your help.
Hope to have any answer from them.
Sggtav