Author Topic: fluke 3330B fault current output  (Read 911 times)
Offline NRP_SANTOSTopic starter
  • Country: pt
« on: December 07, 2021, 02:33:45 pm »
Hello friends, I'm looking for help repairing my fluke 3330b.
his problem is in currents above 3ma. Whichever scale is selected, it always maintains 6ma of output.
Best regards
nrp_santos
Offline alm
Re: fluke 3330B fault current output
« Reply #1 on: December 08, 2021, 12:35:18 pm »
Is it working below 3 mA? Does it jump right to 6 mA as soon as you set the current above 3 mA?
I haven't studied the schematic, but sounds like it might be a bad transistor to me. Although I would also check all voltage rails for ripple (bad caps).
Be aware there are lethal voltages inside, so be very careful operating it with the covers off.
The following users thanked this post: NRP_SANTOS
  • Country: pt
« Reply #2 on: December 09, 2021, 09:44:58 am »
hello Alm,
I appreciate your availability.
It works below 3ma.
above the 3ma range it always maintains the 6ma output value.
best regards,
nrp_santos
Offline Dr. Frank
« Reply #3 on: December 09, 2021, 03:37:37 pm »
Hello,
does this error occur only in constant current mode, or also in constant voltage mode?
Frank
  • Country: pt
« Reply #4 on: December 10, 2021, 09:43:33 am »
Hello Frank,
The problem only occurs in current.
When I select above 3ma.
the signal measured oscilloscope the input and output of the PCBs (A4A2 chopper amp., A4A3 differecial amp., and A4A4 series pass drive) occurs a 100hz oscillation.
NRP_SANTOS
« Last Edit: December 10, 2021, 10:03:58 am by NRP_SANTOS »