Hi
It is quite possible for the 600uV to be the Seebeck effect. As it can be 50/60uV per degree Celsius depending on the materials in contact so you would only need some 10 degrees Celsius temperature difference between ambient temperature and the temperature inside the 7150+. Not difficult for an instrument that has been on power for days.
Have you tried using the rear terminal for your tests?
G Edmonds
The part from the instrument inside to the terminals is made of copper wire with very low thermal EMF. The observed voltage would have to come from the ouside part, starting from the terminals.
50 µV/K is for a materials pair like iron-Konstantan choosen to give an exceptional high voltage. The material with the plugs is more like brass or similar and thus more like in the 10 µV/K range, or even better. The expected error from Seebecke voltage is more like in the 10 µV range.
A stable zero on the current ranges suggests a working ADC and amplifier.
The main candidate are dirty relay contacts - these sometimes in combination with EMI produce some unexpected voltage. With old meters not using the relays for a long time may be a problem, more often than actual contact wear. Cycling the relays a few times can be quite effective (e.g. for a few years) and regular use can keep the contacts clean. It is not uncommon that meters were sitting unused in a shelf for years before beeing sold.
There is usually no much different in the input path betwee the voltage and resistance mode. So its only a few relays or protection elements to check.