I warmed up a 34420a nanovolt meter, and ran some stats on some shorts I made.
The "copper" short is built wit the Lemo nut connector and CdSn solder.
Cheapie #1 and Cheapie #2 are built with Lemo push-pull connectors and regular solder.
All three shorts show similar noise levels.
The copper pins short averages 2nV. The meter is well over a year out of calibration, but this is still pretty good. Or I have a "lucky" short.
The simpler shorts have similar noise levels to the copper short, but show an offset of 22 or 35 nV.
I did not control the number of readings, so these sessions ran several hours each in a non-temperature controlled environment. That likely boosted the spread on the copper connector.
The main takeaways for me are:
The non-copper pin connectors are fine to use in many cases but you must compensate for the offset.
The push-pull connectors are convenient, but you still have to wait for the thermals to settle so you don't get shorter cycles. Plus, the connectors likely take a bit longer to stabilize.
There might be a small thermal offset at the connection pins in the meter.
(the push-pulls are also 1/10 the price of the copper nut connetcors).
Cheapie #1 Copper Cheapie #2
Average (nV) 35 2.0 22.1
SD 2.9 3.7 2.7
P-P 22.7 25.7 19.2
Min 22 -10.9 11.9
Max 44 14.8 31.1
Readings 5109 11824 2449