This is my first post in the EEVblog forum, so here goes...
A few weeks ago, I got an old DSO from a neighbor of mine who said I could have it if I could fix it. He was always going to tear into it and repair it, but that was 10 years ago. According to him, the problem was probably in the vertical and horizontal amps since the calibration signal only took up a small portion of the crt. I turned it on after bringing it home and it functioned just like he said. All of the controls did what they should, the waveform was just very small with some strange artifacts on the leading edge.
I began to read up on the scope with the service manual and began to pull the cover off. I checked each of the supply voltage test points and thought I may have found my problem! The rails are as listed: +6, -6, +12, -12, +100, +170, -60, and -120. All of the probed voltages looked good, except for the 100, 170, and -120. The 170 was about 110V, the -120 was at about -136v, and the most curious was the 100v testing at -1.15v. Upon checking the schematics, each of these voltages appear to be generated from individual taps on a transformer on the power supply board (A7). I could not find any faulty caps with a visual inspection or checking capacitance while in circuit. However, when I probed tap leads 15 and 16, the 100v winding, I found it to overload the ohmeter. The adjacent 170v winding reads 22.4 ohms while still in circuit.
I haven't got much experience diagnosing this complex of equipment so I feel slightly in over my head. Am I looking at a bad transformer, hence a bad board since I cannot find this tranny by itself at all? I would really like to get this scope functioning and make it my personal first oscilliscope!