After the great success I had with my Datron 4000A DC Calibrator, I decided to grab a 4200 so I have an AC source as well.
There appears to be a couple of issues with this unit...
- Firstly, I get a FAIL 2 during self-test.
- During actual use everything is spot on except the 100V range, which is at about 85% of what you would expect.
- After 5 to 10 minutes it trips my house RCD (which the kids are starting to get annoyed about ;-))
I can't find a 4200 manual (of any kind) anywhere, so I'm looking at the 4708 manual which certainly seems to be pretty close, the schematics and board layouts look identical for the key boards, so I suspect they are the same modules.
I'm hoping that the FAIL 2 during self-test is because of the incorrect output on the 100V range, it's not entirely clear in the description of the self-test, it does say FAIL 2 is an over-voltage, and I'm seeing under-voltage.
I have also seen a couple of FAIL 5's when I've been adjusting the output, this is "Safety circuits watchdog" (if the 4708 is the same), but they don't seem to be frequent or easily reproducable.
It looks like the under-voltage is related to the 100V amplifier since this is only in the path in the 100V range, and everything else seems to be ok. I've had a quick look at the supply voltages for it (-400v) and they seem to be ok.
Does anyone have any experience with the 4200? Also, any clues as to why I would be tripping my RCD but only after a delay? I'm assuming it's heat related, but I'm struggling to understand why I'd get heat related earth leakage?
Thanks,
Lee.