Author Topic: tds3000 look what happened  (Read 8532 times)
Offline R_G_B_Topic starter
« on: June 22, 2016, 10:49:55 pm »
Can anyone help with this:
A TDS3000 oscilloscope was hanging when switched on and would just show the boot logo.
I  opened it up and tried disconnecting and reconnecting cables.
I removed the cable from the inverter circuit to the display driver and switch it on hoping, it would go into some kind of diagnostic mode. However when I reconnected the cable and then power up the scope. I got two vertical stripes of the display and no boot display.
See attachments.
Thanks
R_G_B
R_G_B
Re: tds3000 look what happened
« Reply #1 on: June 22, 2016, 10:59:44 pm »
PSU bad caps?
Offline nctnico
« Reply #2 on: June 22, 2016, 11:14:28 pm »
Seems like one or more cables didn't got connected properly. I also don't see why a device would go into diagnostic mode with the inverter cable disconnected  :palm:
« Reply #3 on: June 23, 2016, 04:23:39 am »
See attachments.
Thanks
R_G_B
You haven't given us a lot to go on.
History:
How long have you had it, has the fault just appeared or has there been indications that a problem was developing?
Sure it's a display problem but why, PSU problem (always the first check), display driver, Display connections or software boot problem?
Yeah, so you've done the mandatory fiddle but have you even checked the PSU is to spec, you don't say....and as it's a SMPS you'll probably need another scope. (ripple frequency possibly beyond your DMM range)
Service manual:
http://exodus.poly.edu/~kurt/manuals/manuals/Tektronix/TEK%20TDS%203000%20series%20Service.pdf
I suspect (but not sure) there may be another connector behind the display that interfaces the cable from the Mobo to the LCD panel and it would be best to reseat this one too. I haven't checked the SM but it should show it as a part or in an exploded view.
The following users thanked this post: Jwalling
Offline SL4P
  • There's more value if you figure it out yourself!
« Reply #4 on: June 23, 2016, 05:44:19 am »
The two stripes ojn the display say the backlight inverter seems to be working - so what else happened?
A data ribbon cable, or the inverter output touched something else... ? oops!
Don't ask a question if you aren't willing to listen to the answer.
« Reply #5 on: June 23, 2016, 06:19:35 am »
When the scope is switched on it was just showing the DPO logo and it would just stay there.
So I tried disconnecting and reconnecting cables to try. To see if it was a faulty connection some how and to see if I could force it into a some kind of  diagnostic mode. I disconnected the cable from the display driver and the inverter. And switched on the oscilloscope
And got just a grey screen. I then reconnected the cable and powerd up.
This is all I did. And the display ended up as it is. Seems strange just removing  a cable would do this. Nothing was shorted.
« Last Edit: June 23, 2016, 06:21:26 am by R_G_B_ »
R_G_B
Offline SL4P
« Reply #6 on: June 23, 2016, 06:24:51 am »
I'd be looking <very> closely at the connections and PCB at both ends of the display ribbon cable, and the ribbon itself.
Not much else with what we know.
« Reply #7 on: June 23, 2016, 06:54:55 am »
Yes sorted the cable next to the inverter cable was not seated properly.
Now I need to find out how to get off the stuck DPO boot logo. Can anyone offer any suggestions?
R_G_G
R_G_B
Offline lowimpedance
  • Watts in an ohm?
« Reply #8 on: June 23, 2016, 07:04:09 am »
Not much of a suggestion  but my work DPO4000 series scope failed in the same way a few years back, stuck at the boot screen. Being a work machine apart from checking the power supply was okay I was not going to start poking around the main board. The agent said it was a complete main board swap required and at $5300 for the repair, if it happens again the POS will go straight into landfill.
 good luck
« Reply #9 on: June 23, 2016, 07:55:46 am »
There's a factory adjustment procedure listed on P74 of the manual that is entered during power on boot....could be worth a try to see what's in the menus within.
There might even be an error log there that might point you in the right direction.
Offline AndyC_772
« Reply #10 on: June 23, 2016, 08:33:42 am »
That's standard Tek procedure as far as I know. They quoted me stupid money like that just to change the NVRAM battery on a TDS754D; they wouldn't replace the module alone, but instead would swap out the entire processor board  :palm:
« Reply #11 on: June 23, 2016, 09:49:18 am »
Which factory adjustment manual is this?
Is there a link?
Any other considerations.
The processor seems to run a little hot so it may be stuck in some kind of processing loop?
R_G_B
« Reply #12 on: June 23, 2016, 09:53:52 am »
Is there a link?
Reply #3 Service Manual link
« Reply #13 on: June 23, 2016, 09:56:33 am »
Thanks tautech
R_G_B
R_G_B
« Reply #14 on: June 23, 2016, 01:27:19 pm »
Looks like i managed to force it to boot.
I got a multi meter probe tip and ran it along the ram chips U540 & U550 while it was stuck in a loop and it pulled it out of the loop and the relays clicked and the thing boots.
Next problem cant get anything to work what next and suggestions?
Thanks
R_G_B
« Last Edit: June 23, 2016, 01:50:43 pm by R_G_B_ »
R_G_B
« Reply #15 on: June 23, 2016, 01:29:13 pm »
seems to also be displaying the correct date.
The delay button is stuck on.
R_G_B
R_G_B
« Reply #16 on: June 23, 2016, 02:20:09 pm »
I'm guessing the the ram chips may need to re solderer to the board as it still getting stuck on the display and only when i run the multi meter probe along the chip does it boot?
R_G_B
R_G_B
Offline Anks
    • www.krisanks.wordpress.com
« Reply #17 on: June 23, 2016, 05:30:02 pm »
When you say running the probe along them you mean individually or actually running it along them?
« Reply #18 on: June 23, 2016, 05:36:28 pm »
R_G_B
Offline Jwalling
  • This is work?
« Reply #19 on: June 23, 2016, 06:02:33 pm »
Just hope that there isn't two pins next to each other called power and ground! :-BROKE
Jay
Jay
System error. Strike any user to continue.
« Reply #20 on: June 23, 2016, 06:09:45 pm »
Jay
Poor IC design if it was that way
R_G_B
R_G_B
« Reply #21 on: June 23, 2016, 07:40:39 pm »
R_G_B
That needs be fixed ^^^
Many DSO's use front panel buttons to enter/display states while booting, no surprise at all the boot is stalling.
« Reply #22 on: June 23, 2016, 09:03:18 pm »
It boots the relays click disk drive check and then the front panel flashes.
There is also the 1KHZ voltage on probe compensation.
« Last Edit: June 23, 2016, 09:05:52 pm by R_G_B_ »
R_G_B
« Reply #23 on: June 23, 2016, 09:32:59 pm »
Have you found an error log in the UI?
Is there any clues in that log to point you to the problem?
« Reply #24 on: June 23, 2016, 10:12:40 pm »
I looked at the service manual page 74
I tiried the factory adjustment nothing seems to happen.
R_G_B
R_G_B