Electronics > Repair
TDS3014 adventures
Jarek:
Thank you @sicco. I will do as you write and let you know the results.
Jarek:
I managed to partially repair the oscilloscope. I replaced the processor.
Now it starts correctly.
This has one more fault, on channel 2 there is a sine wave with an amplitude of 100V.
sicco:
A decent sine wave? What frequency? Might be an internal test/calibration signal that stays connected in the ceramic attenuator module?
Check the main board bottom side for R’s and C’s that may have floated away when you applied bottom heat while swapping the CPU.
Jarek:
Thanks for advice .
As usual, you're right. It was a sine wave of about 34 kHz. I washed the PCB again and it's OK now.
soesante:
Hello.. My TDS3012B stop working, after not using it for about a month... there is no trace and diagnostic fail at acquisition ... (the disk drive has been removed so it fail)..I have measure around the voltages and all seems ok.. the 75.75mhz oscl also measure correctly (the 3.3v and the frequency).. is any one could suggest where to start?..i attached some photo of the unit , the fw version , the diagnostic and the screen with no trace... thank you
Navigation
[0] Message Index
[*] Previous page
Go to full version