I'm not sure, I think it's the same PSU that's used in all of the TDS300 and TDS400 series scopes. Apparently the LeCroy 9354 uses an almost identical PSU and I found the schematic for that, but R12 must be different. It says 500 Ohm on the LeCroy schematic but this looks like orange-brown-black, 310 Ohm is not too far out from 500 so maybe it's safe to assume that's what it is.
What I'm shocked to discover is that the chopper transistor tests good. I thought sure that was gonna be dead shorted. Makes me think an insect or lump of moist lint got in there and just started burning rather than a catastrophic failure. There were a couple of other dead bugs inside the scope.