I can't move waveforms horizontally on my TDS520D. The horizontal position knob isn't moving waveforms. I've tried a few troubleshooting steps to no avail:
- Factory Reset (normal and DPO)
- Diag/Err: no failures
- Pressing the help button then turning the horizontal position knob does display its help text
- I replaced the front panel. No change.
- Tried both channels with different probes
- Checked the zoom menu and tried the different lock/unlock functions
Am I missing some hidden locking function which persists through a factory reset? Should I be looking at the acquisition or processing boards? Any advice is greatly appreciated.
-
Did you set the acquisition length to more than 1000 Samples?
Did you turn off the "fit to screen" option?
Did you turn off DPO / InstaVu mode?
-
Did you set the acquisition length to more than 1000 Samples?
Did you turn off the "fit to screen" option?
Did you turn off DPO / InstaVu mode?