Author Topic: Tektronix TDS7254B Signal Path Compensation fail issue and pass all Diagnostic  (Read 5079 times)
Offline snoopy_leiTopic starter
  • Country: cn
Hello everyone,
Would you help me to Close TDS7254B SPC fail issue or give me some suggestions? However thanks for your help in advance.
The TDS7254B can pass all Diagnostic as attach picture and only fail in SPC calibration,however I check the PowerPC board's debug info via few wires to connect PowerPC debug serial port,however the error info as below when run the SPC calibration instruction,
################### topScript ends ###############################
Done executing startup script c:/vxboot/topscript.hw
KERNEL:  RUNNING TIME SLICED AT 10 TICKS PER SLICE!
-> 0x3814820 (aAcqDpoExecutor): 20OCT19 15:38:45 U
0x1641830 (SPC task): cc_loadV: CC_TRIGdcGain3c was 0; clipped
0x1641830 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to 0.000000, failed
0x1641830 (SPC task): cc_loadV: CC_TRIGdcGain4c was 0; clipped
0x1641830 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to 0.000000, failed
total cal time     3.80 minutes
However I have restore the WINXP system with TEK TDS7KB RestoreCD and reinstall the Tekscope software,but the issue same as before,and "M866CompCalibrator 3c Failed" "M866CompCalibrator 4c failed" the info is very confusion,what is M866? It's Chip IC or software instruction? What is 3c and 4c? it's mean the 3rd channel and 4rd Channel?  |O
romPost: I am alive .
uuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuu
romPost: L2 cache turned on.
romPost: MPC107  Walking-one passed.
romPost: MPC107 configuration passed.
romPost: Intel21555 PCITOPCI bridge configuration passed.
romPost: PERFORMING ROM CHECKSUM TEST...
romPost: ROM checksum passed.
romPost: PERFORMING DRAM CELL TEST...
romPost: Enabling data cache...
romPost: Data cache enabled.
romPost: Enabling instruction cache...
romPost: Instruction cache enabled.
romPost: DRAM Cell test passed.
romPost: PERFORMING DRAM MARCH TEST WITH CACHE OFF...
romPost: Disabling data Cache...
romPost: Data cache disabled.
romPost: Disabling instruction cache...
romPost: Instruction cache disabled.
romPost: Writing address to each *address...
romPost: Reading address from each *address...
romPost: DRAM march test passed...
romPost: PERFORMING DRAM MARCH TEST WITH CACHE ON...
romPost: PERFORMING DRAM WALKING-ONE  TEST...
romPost: DRAM Walking-one test passed.
romPost: PERFORMING NV-MEMORY WALKING-ONE  TEST...
romPost: NV-MEMORY Walking-one test passed.
romPost: POWER-UP DIAGNOSTIC COMPLETE.
Hel
                            VxWorks System Boot
Copyright 1984-1996  Wind River Systems, Inc.
CPU: Tektronix HPP - IBM 750CXe
Version: 5.3.1
BSP version: 2.0.1.10
Creation date: Sep 30 2003, 07:53:07
Press any key to stop auto-boot...
 0
auto-booting...
boot device          : sm
processor number     : 0
file name            : c:/vxboot/vxWorks.st
inet on backplane (b): 192.168.0.1:ffffff00
host inet (h)        : 192.168.0.2
gateway inet (g)     : 192.168.0.1
flags (f)            : 0x1008
target name (tn)     : tds7000
startup script (s)   : c:/vxboot/topscript.hw
Initializing backplane net with anchor at 0x1e80000... done.
Backplane anchor at 0x1e80000... Attaching network interface sm0... done
done.
Attaching networPC1 dio device driver installed
k interface lo0... done.
Attaching DIO client... done.
Loading c:/vxboot/vxWorks.st...1773468
Starting at 0x200000...
Host Name: bootHost
User: target
Initializing backplane net with anchor at 0x16e0000... done.
Backplane anchor at 0x16e0000... Attaching network interface sm0... done
done.
Attaching network interface lo0... done.
Adding 6034 symbols for standalone.
      ]]]]]]]]]]]  ]]]]     ]]]]]]]]]]       ]]              ]]]]         (R)
 ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]       Development System
 ]]]]]]]]]]]]]]]]]]]]]]]]]]]       VxWorks version 5.3.1
 ]]]]]]]]]]]]]]]]]]]]]]]]]]       KERNEL: WIND version 2.5
 ]]]]]]]]]]]]]]]]]]]]]]]]]       Copyright Wind River Systems, Inc., 1984-1997
                               CPU: Tektronix HPP - IBM 750CXe.  Processor #0.
                              Memory Size: 0x2000000.  BSP version 2.0.1.10.
                             WDB: Ready.
done
KERNEL: THERMAL CONTROL INITIATED
Executing startup script c:/vxboot/topscript.hw ...
dbFsIoctl: invalid code 3
    ################### topScript starts ###############################
     cd "dbfs:c:/vxBoot"
value = 0 = 0x0
    pDbDev = ramDevCreate(0, 512, 1152, 1152, 0)
new symbol "pDbDev" added to symbol table.
pDbDev = 0x1692d40: value = 23670112 = 0x1692d60 = pDbDev + 0x20
    pDBfs  = dosFsMkfs("/db", pDbDev)
new symbol "pDBfs" added to symbol table.
pDBfs = 0x1691720: value = 23669856 = 0x1692c60
    ld < TekScope.o
value = 23657440 = 0x168fbe0
    ld < TekScopeSM.o
value = 23660544 = 0x1690800
    copy ("TekScope.DBF", "/db/TekScope.DBF")
value = 0 = 0x0
    loggerInit
Serial No: B020790
value = 21822832 = 0x14cfd70
    overrideFPdefaults()
value = 0 = 0x0
    measurePPCtimebase
value = 0 = 0x0
    registerMemPartitionSizes ( 0x00000000 )
value = 0 = 0x0
    semInitForQueueCheckLogMsg
value = 23661184 = 0x1690a80
    initSysWaiter()
value = 0 = 0x0
    raxInit
value = 0 = 0x0
    paiInit
done
value = 16129500 = 0xf61ddc = PaiBaseRegisterTable + 0x804
    configureACL
SUCCESS
value = 0 = 0x0
    verifyAcqBoardID
value = 0 = 0x0
     tcsLoadTableToNVRAM "pc:c:/vxBoot/setup.tcs"
value = 0 = 0x0
     updateAcqProcFirmware("dbfs:c:/vxBoot/acqProc.hex")
0x16987e0 (tShell): AVR FLASH contents match file dbfs:c:/vxBoot/acqProc.hex - no programming needed.
logTask: 1 log messages lost.
value = 1 = 0x1
     cd "pc:"
value = 0 = 0x0
    raxInitializeDevices
done
0x16987e0 (tShell): (Re)starting Acq Processor
value = 23692720 = 0x16985b0
    excHookAdd(stopStressForException)
value = 15840048 = 0xf1b330 = stopStressForException
    dsyInit
value = 0 = 0x0
    dsyWPApplyNextRegs
value = 16110116 = 0xf5d224 = DsyRegisterTable + 0xbd0
    # remove '#' from next line to set powerup scope state to factory default:
    # NvramClearDb = 1
    nvram_initFileSys
value = 21793056 = 0x14c8920
    TmgrInit("/db")
value = 0 = 0x0
    PhxProxyInit
value = 20333344 = 0x1364320
    rtlInit
value =0 = 0x0
    sysExtMemAddToPool memSysPartId,0x02000000,0x1c000000
value = 0 = 0x0
    hwAccount_init
sysEnetAddrGet: 4d:54:5c:c8:c0:64
Scope Model: TDS7254B
value = 0 = 0x0
    GpibInit
value = 502610496 = 0x1df53a40
    wfmMemInit(1)
value = 62205664 = 0x3b52ee0
    vertInstall
value = 62198148 = 0x3b51184
    calConStatus = cc_init(2)
0x16987e0 (tShell): Reading 358 bytes of service cal #1 data from NV[3:0] into 0x3b51000
0x16987e0 (tShell): nvMem_read succeeded
0x16987e0 (tShell): Optical hardware is not present.
calConStatus = 0x127d1dc: value = 0 = 0x0
    vertPostCalRestoreInstall
value = 0 = 0x0
    MathMeasCreateTheMathandMeasurementLibraries
value = 62025376 = 0x3b26ea0
    initAcqMapper
0x16987e0 (tShell): initAcqMapper: Enter
value = 0 = 0x0
    trigMapInit
value = 0 = 0x0
    dsyMapInit
value = 0 = 0x0
    mathMapperInit
value = 0 = 0x0
    utilMapInit
value = 62019520 = 0x3b257c0
    TmgrHookup
value = 0 = 0x0
    PhxProxyHookup
value = 0 = 0x0
    hookupAcqMapper
value = 19321872 = 0x126d410 = AcqMapperRtlDataList
    MathMeasMapHookup
value = 23692552 = 0x1698508
    dsyMapHookup
value = 23692656 = 0x1698570
    utilMapHookup
value = 23692688 = 0x1698590
    rtlHookup
value = 0 = 0x0
    horizMapHookup
value = 23692648 = 0x1698568
    fastFrameMapHookup
value = 23692672 = 0x1698580
    trigMapHookup
value = 23692680 = 0x1698588
    vertMapHookup
    spcMapHookup
    autosetMapHookup
value = 23692712 = 0x16985a8
    timeSyncMapHookup
value = 23692704 = 0x16985a0
    auxTrigMapHookup
    dpoMapHookup
    TmgrStartup
value = 61194384 = 0x3a5c090
    PhxProxyStartup
value = 61143840 = 0x3a4fb20
    taskDelay(2)
value = 0 = 0x0
    gpibHwInitC
value = 0 = 0x0
    gpibTaskSpawn
value = 61017376 = 0x3a30d20
    gpibIoInit "/dbios/DBIO005","/dbios/DBIO006"
value = 60967680 = 0x3a24b00
    enablePowerupDiags = 0
enablePowerupDiags = 0x123dbdc: value = 0 = 0x0
    runPowerupDiags
done
dipSwitch = 0x0, enablePowerupDiags = 0, puMode=0
CalDiagPowerUpDiagsEnabled = 0
!! SKIPPING POWERUP DIAGNOSTICS !!
<< POWERUP DIAGS FINISHED >>
                 execution time   0 sec>>
Test "351" - Acquisition-->HFStep-->PhaseCal started
                 Test execution time 6.3768 sec
value = 0 = 0x0
    rtlStartup
value = 0 = 0x0
    tsComp
value = 0 = 0x0
    paiPrintVersionNumber
Pai version: 4.52
value = 18 = 0x12
DIAGNOSTIC EXECUTION begin
Test "111" - Processor-->Memory-->DRAMWalk1 started
                 Test execution time 6.875e-06 sec
Test "112" - Processor-->Memory-->DRAMCell started
                 Test execution time 0.047812 sec
Test "113" - Processor-->Memory-->DRAMMarch started
                 Test execution time 0.014548 sec
Test "121" - Processor-->PCI Bus-->Scan Test started
                 Test execution time 1.7788e-05 sec
Test "211" - RegAccess-->DACs-->Daculator started
                 Test execution time 0.00036913 sec
Test "221" - RegAccess-->Direct R/W-->DEMUX started
                 Test execution time 0.0023205 sec
Test "222" - RegAccess-->Direct R/W-->Dsy started
                 Test execution time 0.00032442 sec
Test "223" - RegAccess-->Direct R/W-->SBTL started
                 Test execution time 0.00013178 sec
Test "224" - RegAccess-->Direct R/W-->SGTL started
                 Test execution time 0.00029986 sec
Test "231" - RegAccess-->IIC-->Pogo_U1 started
                 Test execution time 0.00040611 sec
Test "241" - RegAccess-->PLL-->Serial started
                 Test execution time 9.2548e-05 sec
Test "251" - RegAccess-->Serial-->AtoD started
                 Test execution time 0.00019014 sec
Test "252" - RegAccess-->Serial-->U1093 started
                 Test execution time 2.7115e-05 sec
Test "253" - RegAccess-->Serial-->U1251 started
                 Test execution time 0.00019577 sec
Test "254" - RegAccess-->Serial-->U1500 started
                 Test execution time 0.00027058 sec
Test "311" - Acquisition-->Memory-->MemData started
                 Test execution time 5.0736 sec
Test "312" - Acquisition-->Memory-->MemAddr started
                 Test execution time 4.9452 sec
Test "313" - Acquisition-->Memory-->DataFormat started
                 Test execution time 2.4118 sec
Test "321" - Acquisition-->Acq Modes-->SubSample started
                 Test execution time 5.3287 sec
Test "322" - Acquisition-->Acq Modes-->PeakDetect started
                 Test execution time 39.147 sec
Test "323" - Acquisition-->Acq Modes-->HiRes started
                 Test execution time 5.3411 sec
Test "331" - Acquisition-->Vertical-->50OhmOvld started
                 Test execution time 0.083668 sec
Test "332" - Acquisition-->Vertical-->ADC Connects started
                 Test execution time 7.0864 sec
Test "341" - Acquisition-->Interrupt-->Bit Tests started
                 Test execution time 4.9321 sec
                 Test execution time 5.8304 sec
Test "361" - Acquisition-->PLL-->Clock Freq started
                 Test execution time 0.44611 sec
Test "371" - Acquisition-->MemSpeed-->DemuxA started
                 Test execution time 3.9665 sec
Test "372" - Acquisition-->MemSpeed-->DemuxB started
                 Test execution time 3.9569 sec
Test "373" - Acquisition-->MemSpeed-->DemuxC started
                 Test execution time 3.9557 sec
Test "374" - Acquisition-->MemSpeed-->DemuxD started
                 Test execution time 3.9562 sec
Test "375" - Acquisition-->MemSpeed-->DemuxE started
                 Test execution time 3.9563 sec
Test "376" - Acquisition-->MemSpeed-->DemuxF started
                 Test execution time 3.9579 sec
Test "377" - Acquisition-->MemSpeed-->DemuxG started
Test "378" - Acquisition-->MemSpeed-->DemuxH started
                 Test execution time 3.9561 sec
Test "411" - Trigger-->Inputs-->Line started
                 Test execution time 0.072346 sec
Test "412" - Trigger-->Inputs-->Ch1 started
                 Test execution time 0.36608 sec
Test "413" - Trigger-->Inputs-->Ch2 started
                 Test execution time 0.35029 sec
Test "414" - Trigger-->Inputs-->Ch3 started
                 Test execution time 0.35027 sec
Test "415" - Trigger-->Inputs-->Ch4 started
                 Test execution time 0.35036 sec
Test "421" - Trigger-->Internals-->TimerA1 started
                 Test execution time 0.10003 sec
Test "422" - Trigger-->Internals-->TimerA2 started
                 Test execution time 0.099752 sec
Test "423" - Trigger-->Internals-->Btrig started
                 Test execution time 0.10234 sec
Test "424" - Trigger-->Internals-->BtrigEvents started
                 Test execution time 0.11884 sec
Test "425" - Trigger-->Internals-->TimeInterp started
                 Test execution time 9.3269e-06 sec
Test "426" - Trigger-->Internals-->Holdoff started
                 Test execution time 0.033352 sec
Test "511" - Misc-->Cal Integrity-->Factory Cal started
                 Test execution time 6.0577e-06 sec
Test "512" - Misc-->Cal Integrity-->SPC (run SPC) started
                 Test execution time 1.8269e-06 sec
Test "521" - Misc-->GPIB-->Interface started
                 Test execution time 0.0005699 sec
DIAGNOSTIC EXECUTION finished in 115.5 sec
« Last Edit: October 21, 2019, 12:22:30 am by snoopy_lei »
  • Country: cn
After I warmed up all night, I ran SPC again and displayed the following message.the "CC_TRIGdcGain3c" constant have some change,but failed same as before,
As we know The DCGain characterized at factory cal and stored along with all other calibration constants in NVRAM,Perhaps it is because the error that has accumulated
over time has caused the current DCGain to exceed the allowable error range.How to adjust the DCgain to the allowable error range? |O Another possibility is that there is
a problem with the NVRAM data |O
-> 0x382ae40 (aAcqExecutor): 27OCT19 01:40:57 U
0x1631bf0 (SPC task): cc_loadV: CC_TRIGdcGain3c was -0.00136806; clipped
0x1631bf0 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to -0.001368, failed
0x1631bf0 (SPC task): cc_loadV: CC_TRIGdcGain4c was -0.00100208; clipped
0x1631bf0 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to -0.001002, failed
0x1594dd0 (SPC task): cc_loadV: CC_TRIGdcGain3c was -0.00115893; clipped
0x1594dd0 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to -0.001159, failed
0x1594dd0 (SPC task): cc_loadV: CC_TRIGdcGain4c was -0.00092366; clipped
0x1594dd0 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to -0.000924, failed
total cal time     3.87 minutes
Offline Perpf
Hi Snoopy!
Obviously you know a lot about the internals of these scopes. Maybe you can give me a hint, where to look for the problem with my defective CSA8000B?
In particular there are two things which could help:
1. Is there a way to read the data from the NVRAMs on PPC and ACQ-Board in order to write them back after replacement?
2. Mine gets stuck during startup of the scope app. PPC shows "P", i get "Kernel running at 10 ticks per slice", then some connection task is going on and somewhere there it gets stuck... Any ideas??
I'd greatly appreciate any help!!!
Offline charlyd
Hi did you ever find the answer i have a csa7404B here with same error on all 4 channels.
Offline mikehank
Hi guys,
The M866 the channel preamp.  Check for any output voltage on the input.  You don’t have a new SPC so there will  be some but not more than 5-6mv
Offline charlyd
@mikehank :  M866 pre amp ....uuh what component are you pointing to i will add my log also to compare ;-)
boot + diag + spc   ( diag gives All Pass)
--------------------------------------------------------------boot ----------------------------------------------------------
Hel
                            VxWorks System Boot
Version: 5.3.1
 0
auto-booting...
boot device          : sm
flags (f)            : 0x1008
done.
Host Name: bootHost
User: target
done.
                             WDB: Ready.
done
     cd "dbfs:c:/vxBoot"
value = 0 = 0x0
    ld < TekScope.o
    ld < TekScopeSM.o
value = 0 = 0x0
    loggerInit
Serial No: B020198
value = 0 = 0x0
    measurePPCtimebase
value = 0 = 0x0
value = 0 = 0x0
    initSysWaiter()
value = 0 = 0x0
    raxInit
value = 0 = 0x0
    paiInit
done
    configureACL
SUCCESS
value = 0 = 0x0
    verifyAcqBoardID
value = 0 = 0x0
value = 0 = 0x0
value = 1 = 0x1
     cd "pc:"
value = 0 = 0x0
done
    dsyInit
value = 0 = 0x0
    dsyWPApplyNextRegs
    # NvramClearDb = 1
    nvram_initFileSys
value = 21792992 = 0x14c88e0
    TmgrInit("/db")
value = 0 = 0x0
    PhxProxyInit
value = 20333280 = 0x13642e0
    rtlInit
value =0 = 0x0
value = 0 = 0x0
    hwAccount_init
sysEnetAddrGet: 1b:15:41:8b:e4:21
Scope Model: CSA7404B
value = 255 = 0xff = DSP_cc_load_pri_byte_at_pri_load_ar1 + 0xd
    GpibInit
value = 502600832 = 0x1df51480
    wfmMemInit(1)
value = 62197472 = 0x3b50ee0
    vertInstall
value = 62189956 = 0x3b4f184
x3b4f000
0x16987e0 (tShell):  Complete PFCal NVRAM initialization.
0x16987e0 (tShell):  Ready to Calibrate Probes.
calConStatus = 0x127d1dc: value = 33263 = 0x81ef
value = 0 = 0x0
value = 61919168 = 0x3b0cfc0
    initAcqMapper
value = 0 = 0x0
    trigMapInit
value = 0 = 0x0
    dsyMapInit
value = 0 = 0x0
    mathMapperInit
value = 0 = 0x0
    utilMapInit
value = 61913312 = 0x3b0b8e0
    TmgrHookup
value = 0 = 0x0
    PhxProxyHookup
value = 0 = 0x0
    hookupAcqMapper
    MathMeasMapHookup
    dsyMapHookup
    utilMapHookup
    rtlHookup
value = 0 = 0x0
    horizMapHookup
    fastFrameMapHookup
    trigMapHookup
    vertMapHookup
    spcMapHookup
    autosetMapHookup
    timeSyncMapHookup
    auxTrigMapHookup
    dpoMapHookup
    TmgrStartup
value = 61080848 = 0x3a40510
    PhxProxyStartup
value = 61030304 = 0x3a33fa0
    taskDelay(2)
value = 0 = 0x0
    gpibHwInitC
value = 0 = 0x0
    gpibTaskSpawn
value = 60904352 = 0x3a153a0
value = 60854656 = 0x3a09180
    runPowerupDiags
done
                 Test execution time 20.7 sec
value = 0 = 0x0
    rtlStartup
value = 0 = 0x0
    tsComp
value = 0 = 0x0
Pai version: 4.52
value = 18 = 0x12
-> 0x3812b40 (aAcqExecutor): 24OCT22 12:28:33 U
--------------------------------------------------------------Diagnostics ----------------------------------------------------------
                 Test execution time 6.8269e-06 sec
                 Test execution time 0.050551 sec
                 Test execution time 0.015118 sec
                 Test execution time 1.7933e-05 sec
                 Test execution time 0.00037678 sec
                 Test execution time 0.0023163 sec
                 Test execution time 0.00032207 sec
                 Test execution time 0.00013216 sec
                 Test execution time 0.00030173 sec
                 Test execution time 0.0003662 sec
                 Test execution time 9.2933e-05 sec
                 Test execution time 0.00019029 sec
                 Test execution time 2.6923e-05 sec
                 Test execution time 0.00019611 sec
                 Test execution time 0.00027024 sec
                 Test execution time 5.0706 sec
                 Test execution time 4.9468 sec
                 Test execution time 5.328 sec
                 Test execution time 5.3412 sec
                 Test execution time 0.083661 sec
                 Test execution time 7.0865 sec
                 Test execution time 4.932 sec
                 Test execution time 18.567 sec
                 Test execution time 0.44619 sec
                 Test execution time 3.9646 sec
                 Test execution time 3.9564 sec
                 Test execution time 3.9565 sec
                 Test execution time 0.074602 sec
                 Test execution time 0.36551 sec
                 Test execution time 0.35025 sec
                 Test execution time 0.3504 sec
                 Test execution time 0.35018 sec
                 Test execution time 0.10015 sec
                 Test execution time 0.10125 sec
                 Test execution time 0.11813 sec
                 Test execution time 0.11882 sec
                 Test execution time 0.033314 sec
                 Test execution time 6.5385e-06 sec
                 Test execution time 3.2212e-06 sec
                 Test execution time 0.00056856 sec
DIAGNOSTIC EXECUTION finished in 128.26 sec
--------------------------------------------------------------SPC  ----------------------------------------------------------
0x381ae30 (SPC task): cc_loadV: CC_TRIGdcGain1b was 0; clipped
0x381ae30 (SPC task): ** M866CompCalibrator 1b failed -- ...wrote cal constant C
C_TRIGdcGain1b to 0.000000, failed
0x381ae30 (SPC task): cc_loadV: CC_TRIGdcGain2b was 0; clipped
0x381ae30 (SPC task): ** M866CompCalibrator 2b failed -- ...wrote cal constant C
C_TRIGdcGain2b to 0.000000, failed
0x381ae30 (SPC task): cc_loadV: CC_TRIGdcGain3b was 0; clipped
0x381ae30 (SPC task): ** M866CompCalibrator 3b failed -- ...wrote cal constant C
C_TRIGdcGain3b to 0.000000, failed
0x381ae30 (SPC task): cc_loadV: CC_TRIGdcGain4b was 0; clipped
0x381ae30 (SPC task): ** M866CompCalibrator 4b failed -- ...wrote cal constant C
C_TRIGdcGain4b to 0.000000, failed
total cal time     4.40 minutes
« Last Edit: October 24, 2022, 02:44:48 pm by charlyd »
Offline mikehank
So your failure for spc is common.  Diags are will not pick up the M866 calibration failures.  What is happening is spc is applying a calibration signal to the preamp for various measurements and the signal is out of range(too small or too large) and fails.  Attached are a couple of pic's that may help.  Install a TNC-BNC to the input of a channel and measure for mV.  You should not see very much under 5-6 mv or so.  If you have more this is the source of the out of range M866 failure.
Offline mikehank
Here is another file
Offline charlyd
Hi Mikehank, yes i have seen this picture but you think the problem is pre-amp related? ok but for all channels then?
does any body have  a picture of those pre-amp without the heatsink.  so the code on top of the chips.
ah i just saw you posted 2 messages (with xlsx)...thanks for your help and support...
« Last Edit: January 07, 2023, 05:40:14 pm by charlyd »
Offline matthuszagh
charlyd, did you ever figure this out? I'm seeing something similar.
Offline charlyd
hi Matthus , no i parked the unit and went to the next one.
So, you have a similiar issue ?
btw somebody explain when 351 - Acquisition-->HFStep-->PhaseCal is displayed  in Diagnostic.
« Last Edit: November 12, 2023, 08:08:28 pm by charlyd »
Offline charlyd
hi is there any clearance on the M866 issue
M866CompCal failed  i now have a TDS7404B  with  ACQ Diags all pass.
this 7404B unit also has a ch3 and ch4 error exact like the 7254B in first post.  it looks like it should be a component or circuit for two channels. So that there are two of the same circuits/components for each two channels and one of the two circuits is failing.
so i guess there must be a relation. It is a bit difficult to dive deeper because you need to measure on the back side of the ACQ board.  someone already experienced with let say extension cables or something?
specially to jaywalling because of your experience
any m866 news?
i am willing to pay $200 to whom who can point me to the M866 Comp Cal failed solution. like in first post
« Last Edit: February 14, 2024, 10:08:32 pm by charlyd »
Offline charlyd
Hi its a while since....but i was stuck searching....
Today my eye caught this picture... not saying that this is found on the CSA7404B series.
but i am still curious if somebody has a failed "B series ACQ board" and is willing to pull of some heatsinks...
to get a better board layout... searching for the mysterious M866 in the B series 7000 scopes
exact board picture see post #7 above
because i have seen only ch 3 and ch4 failing and like in my case all channels failing. This would suggest there are two (or more but i doubt) of these chips equipt.
« Last Edit: December 07, 2024, 01:23:28 pm by charlyd »