| snoopy_lei:
Hello everyone, Would you help me to Close TDS7254B SPC fail issue or give me some suggestions? However thanks for your help in advance. The TDS7254B can pass all Diagnostic as attach picture and only fail in SPC calibration,however I check the PowerPC board's debug info via few wires to connect PowerPC debug serial port,however the error info as below when run the SPC calibration instruction, --- Code: ---################### topScript ends ############################### Done executing startup script c:/vxboot/topscript.hw KERNEL: RUNNING TIME SLICED AT 10 TICKS PER SLICE! -> 0x3814820 (aAcqDpoExecutor): 20OCT19 15:38:45 U 0x1641830 (SPC task): cc_loadV: CC_TRIGdcGain3c was 0; clipped 0x1641830 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to 0.000000, failed . ** 0x1641830 (SPC task): cc_loadV: CC_TRIGdcGain4c was 0; clipped 0x1641830 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to 0.000000, failed . ** total cal time 3.80 minutes --- End code --- However I have restore the WINXP system with TEK TDS7KB RestoreCD and reinstall the Tekscope software,but the issue same as before,and "M866CompCalibrator 3c Failed" "M866CompCalibrator 4c failed" the info is very confusion,what is M866? It's Chip IC or software instruction? What is 3c and 4c? it's mean the 3rd channel and 4rd Channel? |O --- Code: ---romPost: I am alive . uuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuuu romPost: L2 cache turned on. romPost: MPC107 Walking-one passed. romPost: MPC107 configuration passed. romPost: Intel21555 PCITOPCI bridge configuration passed. romPost: PERFORMING ROM CHECKSUM TEST... romPost: ROM checksum passed. romPost: PERFORMING DRAM CELL TEST... romPost: Enabling data cache... romPost: Data cache enabled. romPost: Enabling instruction cache... romPost: Instruction cache enabled. romPost: DRAM Cell test passed. romPost: PERFORMING DRAM MARCH TEST WITH CACHE OFF... romPost: Disabling data Cache... romPost: Data cache disabled. romPost: Disabling instruction cache... romPost: Instruction cache disabled. romPost: Writing address to each *address... romPost: Reading address from each *address... romPost: DRAM march test passed... romPost: PERFORMING DRAM MARCH TEST WITH CACHE ON... romPost: Enabling data cache... romPost: Data cache enabled. romPost: Enabling instruction cache... romPost: Instruction cache enabled. romPost: Writing address to each *address... romPost: Reading address from each *address... romPost: DRAM march test passed... romPost: PERFORMING DRAM WALKING-ONE TEST... romPost: DRAM Walking-one test passed. romPost: PERFORMING NV-MEMORY WALKING-ONE TEST... romPost: NV-MEMORY Walking-one test passed. romPost: Disabling data Cache... romPost: Data cache disabled. romPost: Disabling instruction cache... romPost: Instruction cache disabled. romPost: Enabling instruction cache... romPost: Instruction cache enabled. romPost: POWER-UP DIAGNOSTIC COMPLETE. Hel VxWorks System Boot Copyright 1984-1996 Wind River Systems, Inc. CPU: Tektronix HPP - IBM 750CXe Version: 5.3.1 BSP version: 2.0.1.10 Creation date: Sep 30 2003, 07:53:07 Press any key to stop auto-boot... 0 auto-booting... boot device : sm processor number : 0 file name : c:/vxboot/vxWorks.st inet on backplane (b): 192.168.0.1:ffffff00 host inet (h) : 192.168.0.2 gateway inet (g) : 192.168.0.1 flags (f) : 0x1008 target name (tn) : tds7000 startup script (s) : c:/vxboot/topscript.hw Initializing backplane net with anchor at 0x1e80000... done. Backplane anchor at 0x1e80000... Attaching network interface sm0... done done. Attaching networPC1 dio device driver installed k interface lo0... done. Attaching DIO client... done. Loading c:/vxboot/vxWorks.st...1773468 Starting at 0x200000... Host Name: bootHost User: target Initializing backplane net with anchor at 0x16e0000... done. Backplane anchor at 0x16e0000... Attaching network interface sm0... done done. PC1 dio device driver installed Attaching network interface lo0... done. Adding 6034 symbols for standalone. ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]] ]]]] ]]]]]]]]]] ]] ]]]] (R) ] ]]]]]]]]] ]]]]]] ]]]]]]]] ]] ]]]] ]] ]]]]]]] ]]]]]]]] ]]]]]] ] ]] ]]]] ]]] ]]]]] ] ]]] ] ]]]] ]]] ]]]]]]]]] ]]]] ]] ]]]] ]] ]]]]] ]]]] ]]] ]] ] ]]] ]] ]]]]] ]]]]]] ]] ]]]]]]] ]]]] ]] ]]]] ]]]]] ] ]]]] ]]]]] ]]]]]]]] ]]]] ]] ]]]] ]]]]]]] ]]]] ]]]]]] ]]]]] ]]]]]] ] ]]]]] ]]]] ]] ]]]] ]]]]]]]] ]]]] ]]]]]]] ]]]]] ] ]]]]]] ] ]]] ]]]] ]] ]]]] ]]]] ]]]] ]]]] ]]]]]]]] ]]]]] ]]] ]]]]]]] ] ]]]]]]] ]]]] ]]]] ]]]] ]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]] Development System ]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]] VxWorks version 5.3.1 ]]]]]]]]]]]]]]]]]]]]]]]]]] KERNEL: WIND version 2.5 ]]]]]]]]]]]]]]]]]]]]]]]]] Copyright Wind River Systems, Inc., 1984-1997 CPU: Tektronix HPP - IBM 750CXe. Processor #0. Memory Size: 0x2000000. BSP version 2.0.1.10. WDB: Ready. done KERNEL: THERMAL CONTROL INITIATED Executing startup script c:/vxboot/topscript.hw ... dbFsIoctl: invalid code 3 ################### topScript starts ############################### cd "dbfs:c:/vxBoot" value = 0 = 0x0 pDbDev = ramDevCreate(0, 512, 1152, 1152, 0) new symbol "pDbDev" added to symbol table. pDbDev = 0x1692d40: value = 23670112 = 0x1692d60 = pDbDev + 0x20 pDBfs = dosFsMkfs("/db", pDbDev) new symbol "pDBfs" added to symbol table. pDBfs = 0x1691720: value = 23669856 = 0x1692c60 ld < TekScope.o value = 23657440 = 0x168fbe0 ld < TekScopeSM.o value = 23660544 = 0x1690800 copy ("TekScope.DBF", "/db/TekScope.DBF") value = 0 = 0x0 loggerInit Serial No: B020790 value = 21822832 = 0x14cfd70 overrideFPdefaults() value = 0 = 0x0 measurePPCtimebase value = 0 = 0x0 registerMemPartitionSizes ( 0x00000000 ) value = 0 = 0x0 semInitForQueueCheckLogMsg value = 23661184 = 0x1690a80 initSysWaiter() value = 0 = 0x0 raxInit value = 0 = 0x0 paiInit done value = 16129500 = 0xf61ddc = PaiBaseRegisterTable + 0x804 configureACL SUCCESS value = 0 = 0x0 verifyAcqBoardID value = 0 = 0x0 tcsLoadTableToNVRAM "pc:c:/vxBoot/setup.tcs" value = 0 = 0x0 updateAcqProcFirmware("dbfs:c:/vxBoot/acqProc.hex") 0x16987e0 (tShell): AVR FLASH contents match file dbfs:c:/vxBoot/acqProc.hex - no programming needed. logTask: 1 log messages lost. value = 1 = 0x1 cd "pc:" value = 0 = 0x0 raxInitializeDevices done 0x16987e0 (tShell): (Re)starting Acq Processor value = 23692720 = 0x16985b0 excHookAdd(stopStressForException) value = 15840048 = 0xf1b330 = stopStressForException dsyInit value = 0 = 0x0 dsyWPApplyNextRegs value = 16110116 = 0xf5d224 = DsyRegisterTable + 0xbd0 # remove '#' from next line to set powerup scope state to factory default: # NvramClearDb = 1 nvram_initFileSys value = 21793056 = 0x14c8920 TmgrInit("/db") value = 0 = 0x0 PhxProxyInit value = 20333344 = 0x1364320 rtlInit value =0 = 0x0 sysExtMemAddToPool memSysPartId,0x02000000,0x1c000000 value = 0 = 0x0 hwAccount_init sysEnetAddrGet: 4d:54:5c:c8:c0:64 Scope Model: TDS7254B value = 0 = 0x0 GpibInit value = 502610496 = 0x1df53a40 wfmMemInit(1) value = 62205664 = 0x3b52ee0 vertInstall value = 62198148 = 0x3b51184 calConStatus = cc_init(2) 0x16987e0 (tShell): Reading 358 bytes of service cal #1 data from NV[3:0] into 0x3b51000 0x16987e0 (tShell): nvMem_read succeeded 0x16987e0 (tShell): Optical hardware is not present. calConStatus = 0x127d1dc: value = 0 = 0x0 vertPostCalRestoreInstall value = 0 = 0x0 MathMeasCreateTheMathandMeasurementLibraries value = 62025376 = 0x3b26ea0 initAcqMapper 0x16987e0 (tShell): initAcqMapper: Enter value = 0 = 0x0 trigMapInit value = 0 = 0x0 dsyMapInit value = 0 = 0x0 mathMapperInit value = 0 = 0x0 utilMapInit value = 62019520 = 0x3b257c0 TmgrHookup value = 0 = 0x0 PhxProxyHookup value = 0 = 0x0 hookupAcqMapper value = 19321872 = 0x126d410 = AcqMapperRtlDataList MathMeasMapHookup value = 23692552 = 0x1698508 dsyMapHookup value = 23692656 = 0x1698570 utilMapHookup value = 23692688 = 0x1698590 rtlHookup value = 0 = 0x0 horizMapHookup value = 23692648 = 0x1698568 fastFrameMapHookup value = 23692672 = 0x1698580 trigMapHookup value = 23692680 = 0x1698588 vertMapHookup value = 23692672 = 0x1698580 spcMapHookup value = 23692672 = 0x1698580 autosetMapHookup value = 23692712 = 0x16985a8 timeSyncMapHookup value = 23692704 = 0x16985a0 auxTrigMapHookup value = 23692688 = 0x1698590 dpoMapHookup value = 23692680 = 0x1698588 TmgrStartup value = 61194384 = 0x3a5c090 PhxProxyStartup value = 61143840 = 0x3a4fb20 taskDelay(2) value = 0 = 0x0 gpibHwInitC value = 0 = 0x0 gpibTaskSpawn value = 61017376 = 0x3a30d20 gpibIoInit "/dbios/DBIO005","/dbios/DBIO006" value = 60967680 = 0x3a24b00 enablePowerupDiags = 0 enablePowerupDiags = 0x123dbdc: value = 0 = 0x0 runPowerupDiags done dipSwitch = 0x0, enablePowerupDiags = 0, puMode=0 CalDiagPowerUpDiagsEnabled = 0 !! SKIPPING POWERUP DIAGNOSTICS !! << POWERUP DIAGS FINISHED >> execution time 0 sec>> Test "351" - Acquisition-->HFStep-->PhaseCal started Test execution time 6.3768 sec value = 0 = 0x0 rtlStartup value = 0 = 0x0 tsComp value = 0 = 0x0 paiPrintVersionNumber Pai version: 4.52 value = 18 = 0x12 ################### topScript ends ############################### Done executing startup script c:/vxboot/topscript.hw KERNEL: RUNNING TIME SLICED AT 10 TICKS PER SLICE! -> 0x3814820 (aAcqDpoExecutor): 20OCT19 15:38:45 U 0x1641830 (SPC task): cc_loadV: CC_TRIGdcGain3c was 0; clipped 0x1641830 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to 0.000000, failed . ** 0x1641830 (SPC task): cc_loadV: CC_TRIGdcGain4c was 0; clipped 0x1641830 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to 0.000000, failed . ** total cal time 3.80 minutes DIAGNOSTIC EXECUTION begin Test "111" - Processor-->Memory-->DRAMWalk1 started Test execution time 6.875e-06 sec Test "112" - Processor-->Memory-->DRAMCell started Test execution time 0.047812 sec Test "113" - Processor-->Memory-->DRAMMarch started Test execution time 0.014548 sec Test "121" - Processor-->PCI Bus-->Scan Test started Test execution time 1.7788e-05 sec Test "211" - RegAccess-->DACs-->Daculator started Test execution time 0.00036913 sec Test "221" - RegAccess-->Direct R/W-->DEMUX started Test execution time 0.0023205 sec Test "222" - RegAccess-->Direct R/W-->Dsy started Test execution time 0.00032442 sec Test "223" - RegAccess-->Direct R/W-->SBTL started Test execution time 0.00013178 sec Test "224" - RegAccess-->Direct R/W-->SGTL started Test execution time 0.00029986 sec Test "231" - RegAccess-->IIC-->Pogo_U1 started Test execution time 0.00040611 sec Test "241" - RegAccess-->PLL-->Serial started Test execution time 9.2548e-05 sec Test "251" - RegAccess-->Serial-->AtoD started Test execution time 0.00019014 sec Test "252" - RegAccess-->Serial-->U1093 started Test execution time 2.7115e-05 sec Test "253" - RegAccess-->Serial-->U1251 started Test execution time 0.00019577 sec Test "254" - RegAccess-->Serial-->U1500 started Test execution time 0.00027058 sec Test "311" - Acquisition-->Memory-->MemData started Test execution time 5.0736 sec Test "312" - Acquisition-->Memory-->MemAddr started Test execution time 4.9452 sec Test "313" - Acquisition-->Memory-->DataFormat started Test execution time 2.4118 sec Test "321" - Acquisition-->Acq Modes-->SubSample started Test execution time 5.3287 sec Test "322" - Acquisition-->Acq Modes-->PeakDetect started Test execution time 39.147 sec Test "323" - Acquisition-->Acq Modes-->HiRes started Test execution time 5.3411 sec Test "331" - Acquisition-->Vertical-->50OhmOvld started Test execution time 0.083668 sec Test "332" - Acquisition-->Vertical-->ADC Connects started Test execution time 7.0864 sec Test "341" - Acquisition-->Interrupt-->Bit Tests started Test execution time 4.9321 sec Test "351" - Acquisition-->HFStep-->PhaseCal started Test execution time 5.8304 sec Test "361" - Acquisition-->PLL-->Clock Freq started Test execution time 0.44611 sec Test "371" - Acquisition-->MemSpeed-->DemuxA started Test execution time 3.9665 sec Test "372" - Acquisition-->MemSpeed-->DemuxB started Test execution time 3.9569 sec Test "373" - Acquisition-->MemSpeed-->DemuxC started Test execution time 3.9557 sec Test "374" - Acquisition-->MemSpeed-->DemuxD started Test execution time 3.9562 sec Test "375" - Acquisition-->MemSpeed-->DemuxE started Test execution time 3.9563 sec Test "376" - Acquisition-->MemSpeed-->DemuxF started Test execution time 3.9579 sec Test "377" - Acquisition-->MemSpeed-->DemuxG started Test execution time 3.9557 sec Test "378" - Acquisition-->MemSpeed-->DemuxH started Test execution time 3.9561 sec Test "411" - Trigger-->Inputs-->Line started Test execution time 0.072346 sec Test "412" - Trigger-->Inputs-->Ch1 started Test execution time 0.36608 sec Test "413" - Trigger-->Inputs-->Ch2 started Test execution time 0.35029 sec Test "414" - Trigger-->Inputs-->Ch3 started Test execution time 0.35027 sec Test "415" - Trigger-->Inputs-->Ch4 started Test execution time 0.35036 sec Test "421" - Trigger-->Internals-->TimerA1 started Test execution time 0.10003 sec Test "422" - Trigger-->Internals-->TimerA2 started Test execution time 0.099752 sec Test "423" - Trigger-->Internals-->Btrig started Test execution time 0.10234 sec Test "424" - Trigger-->Internals-->BtrigEvents started Test execution time 0.11884 sec Test "425" - Trigger-->Internals-->TimeInterp started Test execution time 9.3269e-06 sec Test "426" - Trigger-->Internals-->Holdoff started Test execution time 0.033352 sec Test "511" - Misc-->Cal Integrity-->Factory Cal started Test execution time 6.0577e-06 sec Test "512" - Misc-->Cal Integrity-->SPC (run SPC) started Test execution time 1.8269e-06 sec Test "521" - Misc-->GPIB-->Interface started Test execution time 0.0005699 sec DIAGNOSTIC EXECUTION finished in 115.5 sec --- End code --- |
| snoopy_lei:
After I warmed up all night, I ran SPC again and displayed the following message.the "CC_TRIGdcGain3c" constant have some change,but failed same as before, As we know The DCGain characterized at factory cal and stored along with all other calibration constants in NVRAM,Perhaps it is because the error that has accumulated over time has caused the current DCGain to exceed the allowable error range.How to adjust the DCgain to the allowable error range? |O Another possibility is that there is a problem with the NVRAM data |O --- Code: --- ################### topScript ends ############################### Done executing startup script c:/vxboot/topscript.hw KERNEL: RUNNING TIME SLICED AT 10 TICKS PER SLICE! -> 0x382ae40 (aAcqExecutor): 27OCT19 01:40:57 U 0x1631bf0 (SPC task): cc_loadV: CC_TRIGdcGain3c was -0.00136806; clipped 0x1631bf0 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to -0.001368, failed . ** 0x1631bf0 (SPC task): cc_loadV: CC_TRIGdcGain4c was -0.00100208; clipped 0x1631bf0 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to -0.001002, failed . ** total cal time 3.80 minutes 0x1594dd0 (SPC task): cc_loadV: CC_TRIGdcGain3c was -0.00115893; clipped 0x1594dd0 (SPC task): ** M866CompCalibrator 3c failed -- ...wrote cal constant CC_TRIGdcGain3c to -0.001159, failed . ** 0x1594dd0 (SPC task): cc_loadV: CC_TRIGdcGain4c was -0.00092366; clipped 0x1594dd0 (SPC task): ** M866CompCalibrator 4c failed -- ...wrote cal constant CC_TRIGdcGain4c to -0.000924, failed . ** total cal time 3.87 minutes --- End code --- |
| Perpf:
Hi Snoopy! Obviously you know a lot about the internals of these scopes. Maybe you can give me a hint, where to look for the problem with my defective CSA8000B? In particular there are two things which could help: 1. Is there a way to read the data from the NVRAMs on PPC and ACQ-Board in order to write them back after replacement? 2. Mine gets stuck during startup of the scope app. PPC shows "P", i get "Kernel running at 10 ticks per slice", then some connection task is going on and somewhere there it gets stuck... Any ideas?? I'd greatly appreciate any help!!! |
| charlyd:
Hi did you ever find the answer i have a csa7404B here with same error on all 4 channels. |
| mikehank:
Hi guys, The M866 the channel preamp. Check for any output voltage on the input. You don’t have a new SPC so there will be some but not more than 5-6mv |