Author Topic: tek 2215  (Read 7784 times)
Offline m3vuvTopic starter
tek 2215
« on: October 20, 2021, 03:33:07 pm »
Hi all,i have a tek 2215 just delivered from ebay in a crusty condition,i want to get inside of it to clean it,trouble is the service manual says remove the 2 screws holding the rear cover,then slide the cover off,i have done this but it seems solid!,am i missing something here,i dont want to break anything,tia paul m3vuv.
Re: tek 2215
« Reply #1 on: October 20, 2021, 03:53:06 pm »
dont worry about this,i have the covers off,there was a srew in one side i haddnt spotted,while im in there are there any caps that i should change,ie riffa etc?.
Offline bob91343
Re: tek 2215
« Reply #2 on: October 20, 2021, 05:23:56 pm »
If it's not brokem, I wouldn't fix it.
Offline David Hess
  • DavidH
Re: tek 2215
« Reply #3 on: October 21, 2021, 12:36:09 am »
Before you continue, double check whether it is a 2215 or 2215A; they are very different instruments, and the service manuals are not interchangeable.
Re: tek 2215
« Reply #4 on: October 21, 2021, 02:40:13 am »
it isnt the a version,im more concerned with doing some preventitive maintenance while i have it apart realy.
Re: tek 2215
« Reply #5 on: October 21, 2021, 03:10:14 am »
just noticed both channel 1 and 2 are way to low down the screen,as in when fed a sine wave etc the lower portion of the trace is off screen,the position knobs dont seen to do a lot,any ideas where to look,it does it on both channels.
Offline David Hess
  • DavidH
Re: tek 2215
« Reply #6 on: October 21, 2021, 03:48:42 am »
Is the amplitude correct?
The output load resistors on the vertical CRT amplifier are known to overheat and go open on the 2213/2215 which might cause that.
Re: tek 2215
« Reply #7 on: October 21, 2021, 04:05:02 am »
its hard to tell on the amplitude as i cant see the bottom of the trace,any idea where the resistors are located ?.
Re: tek 2215
« Reply #8 on: October 21, 2021, 04:23:20 am »
if the load resistors are r378,r379,r389 and r 388 they measure as follows r378=314 ohm,r379=309 ohm,r389=321 ohm and r388=316 ohm.
Re: tek 2215
« Reply #9 on: October 21, 2021, 04:23:52 am »
none show signs of getting hot.
Offline andy3055
Re: tek 2215
« Reply #10 on: October 21, 2021, 05:35:34 am »
I have a 2215 and will follow this thread.
Re: tek 2215
« Reply #11 on: October 21, 2021, 05:48:44 am »
hi andy,im pretty sure its q386 and q 387 in the vert amp that are shot,im trying to find out if i can sub bfy 50 or bc108 transistors in there place as i have these to hand
Re: tek 2215
« Reply #12 on: October 21, 2021, 06:07:40 am »
testing q 376 in circuit,as in diode voltage drop,+ve probe on the base i get base to emitter 0.1v and base to collector 0.7v drops,the base to emitter junction looks iffy to me.
Re: tek 2215
« Reply #13 on: October 21, 2021, 03:32:31 pm »
does the transistor readings make it leaky or low gain,ie,should i replace it?,tia m3vuv.
Offline andy3055
Re: tek 2215
« Reply #14 on: October 21, 2021, 04:22:15 pm »
Q386 and 387 are listed as BFR96 and S006075 respectively. I could not find any data so far on these. I would check the voltages as per the service manual first. Also, if anything is amiss, check them out of the circuit.
Noticed your last post says Q376. Is that a mistake?
Re: tek 2215
« Reply #15 on: October 21, 2021, 04:28:51 pm »
sorry meant to say q386,anyway i changed it for a bfy50,made no difference,also the two resistors feeding the deflection plates should be 51 ohm,mine both read 57,would that make much difference?
Offline andy3055
Re: tek 2215
« Reply #16 on: October 21, 2021, 06:51:56 pm »
I am not sure. although I don't think it matters, you can eliminate by putting in new parts there. I have not gone into details very much on this scope as mine is working with no issue. If I get some time to read the manual, I will make some comments here.
Offline andy3055
Re: tek 2215
« Reply #17 on: October 21, 2021, 07:09:59 pm »
CH1/CH2 position controls are associated with Q177/187 and 277/287, respectively. Since both traces are off to one side, I would start tracing backwards from those. I think it could be a voltage related to those and/or the transistors before them.
I am looking at "CH1 &CH2 Vertical Preamps" - foldout 2.
Offline David Hess
  • DavidH
Re: tek 2215
« Reply #18 on: October 22, 2021, 01:07:15 am »
That is odd.  They should show signs of running hot and the printed circuit board should be discolored under them.
Re: tek 2215
« Reply #19 on: October 22, 2021, 03:17:53 am »
well,i removed q387,it tests ok bith a beta of 84.
Re: tek 2215
« Reply #20 on: October 22, 2021, 03:19:56 am »
what page of the manual are q 177 ect on?,i cant find it.
Offline andy3055
Re: tek 2215
« Reply #21 on: October 22, 2021, 04:13:37 am »
Foldout 2. Check and Check vertical preamps. You also got some good feedback on Tek groups.
Re: tek 2215
« Reply #22 on: October 22, 2021, 05:38:39 am »
i only have the service manual on pdf,ive just replaced the 4 riffa caps in the psu,used blue x types,only 2.1 nf not 2.2 like the originals but rated at 8kv.
Offline andy3055
Re: tek 2215
« Reply #23 on: October 22, 2021, 06:14:34 am »
Open the pdf. Do a ctrl+f and that will open a little field for search. Type in Q177 and see if it will find it. I can check the pdf and get back to you in a bit.
Re: tek 2215
« Reply #24 on: October 22, 2021, 06:22:14 am »
the search cant find it.