Hi Folks,
Im hoping the forum can help me with a repair project ive been struggling with, which i cannot seem to solve. I have a Tektronix 2225 50MHz BW CRT Scope. Just like the one which featured in one of Daves videos. It has two issues which i believe are linked, but am not completely sure. There are two conditions i have noticed;
1) With no probes connected, with AC or DC coupling on a selected input (say CH1) on 5mV/div with the x10 Y (giving 500uV/div resolution) i am measuring 18KHz noise. Does it on both CH1 and CH2, is in perfect phase with the PSU switching cct too.
2) With the Input coupling to GND, on 50nS/Div (i.e the fastest timebase) to 0.2uS/div there is distortion at the start of the CRT beam. With the input grounded, the amplitude of the "squiggle" stays the same on all Volt/Division scales, constant with X or Y adjustments. It is "stretched" when the X axis is set to x5, x10 etc.
I had actually noticed symptom 2 first (the squiggle). Given it was constant no matter the voltage division, i didnt think it was between the Input or the attenuator, so started at the Output of the Final Y amp and traced back from the Y Final amp. I found the attached waveform to the input of the Y Final amp stage from the delay line as well as the input of the "Delay line driver circuit". I didnt bother to trace it back further, because when i started probing on the +8.6/-8.6V rails, i found the noise was there too. This squiggle is visible on all input signals, such as a 20MHz sinewave input, you can see this signal superimposed onto the Sinewave, this is both visible on the CRT and when probing the Y input with my other scope. When changing time base to say 0.1uS or 0.2uS, you can see the squiggle get condensed, which says to me it could be occurring at the same rate as the sweep rate of the scope.
I assumed that this noise was coming from the PSU, this was supported by a few online posts that these Scopes do suffer from Switch mode noise so a common approach is to "replace all the electrolytic caps" so i did just that. All Caps within the PSU Area of the board, were replaced with the best quality caps i can get, the lowest ESR etc.
After this i measured and adjusted the -8.6V rail as per the service manual,
-8.6 = -8.6103
+5.1 = 5.1087
+8.6 = 8.7646
+38 = 37.606
+100 = +97.917
Now while all of these DC readings are within their respective limits. The ripple/Vpk to pk readings are not so. The +100 and +38V rails are in spec, but the +/-8.6 and +5.1V are all out of limits on the noise. I did my best to measure the noise with a x1 Probe (as per the service manual) and a short scope spring for the reference (avoiding the croc clip on the probe). The "typical" noise is in spec, but there are higher voltage spikes occurring on the rail
After changing the PSU Caps, the noise on the display was no better either.
The PSU inverter, which is based on a TL594, and drives a push pull inverter using a number of BJT devices, I probed the bases of Q930/Q960 and while the Service manual shows a trapezoidal waveform (imagine a triangle wave with flat spots on the peak/trough), mine was "skewed" and had some ringing on the rising and falling edges (will try and get a scope capture for a follow up). However the ringing didnt quite match up with the "squiggles" on the display. However, the noise measured on the 500uV/Div resolution (symptom No.1), was perfectly triggered to that of the Inverter switching.
Incase it was switching noise from a low line voltage, i did put the scope on my Variac and bumped it up to just over 240VAC (was a typical 230VAC during previous testing) and the noise was the same with no changes at all.
Im a stumped as to where to really go next, the main bulk caps are now replaced, while there are some 15uF caps dotted around the board which i have not replaced, I cannot actually find anything wrong with the scope, but it has a seemingly very noisy PSU.
As the 2nd symptom appears to be in sync with the Sweep rate, ill do some digging around that circuit, and get traces of the noise i am measuring on the PSU rails and will feed back any progress.
Id really appreciate any help with this, marginal or noise issues like this are my biggest weakness when fault finding, ive spent months on/off with this and i really wish i could solve it.
Thanks,
Taylor