Electronics => Repair => Topic started by: dsw8191 on February 27, 2020, 07:48:24 pm
-
I've acquired a Tektronix 2236. It works but there is a distortion in the traces even when the inputs are grounded (see IMG_1095). The degree of distortion is dependent on Y position - worse when the trace is above mid CRT. Both channel traces superimpose when moved to the same Y position. The X position of the distortion is independent of time base setting.
I've searched the forum for a similar problem without success. Hope someone wiser than I can provide an explanation for this.
-
My vote, CRT is damaged beyond repair, possibly the scope has been dropped.
Good news, the X and Y electronics are fine.
More advice if you scroll half way down this page...
https://www.sphere.bc.ca/test/tek-parts/tek-info.html (https://www.sphere.bc.ca/test/tek-parts/tek-info.html)
and another one...
https://www.eevblog.com/forum/repair/tek-2211-crt-weirdness/ (https://www.eevblog.com/forum/repair/tek-2211-crt-weirdness/)